Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
#59805 19/01/12 2:41 PM
Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Dear all
I am having problem in one of SLE2000 infant ventilator.
I have calibrated it electronically & phneumatically.
Problem is that it is showing PIP HIGH alarm continously.
Also by changing the alarm setting of PIP control the ventilator is making unnecessary PEEP of approx 10cmH2O..
Can some one help me out here what i should do & look for.
Waiting for quick reply.
regards

Joined: Nov 2010
Posts: 164
Mentor
Offline
Mentor
Joined: Nov 2010
Posts: 164
in what mode are you working pal? as far as i know, PIP is used in Support ventilation mode and PEEP is in Assist /Control Mode...


There is no work worth dying for.
Joined: Oct 2007
Posts: 75
Adept
Offline
Adept
Joined: Oct 2007
Posts: 75
Not sure what you are asking, what mode are you in?
you have to remember with this vent you set up the CPAP/PEEP first with the pressure display set to MIN, i would suggest in cpap mode, then you set up PIP with pressure display set to max in CMV is best mode. Remember after setup, if PEEP pressure is adjusted this the will be reflecte on you PIP so if you have a Peep of 4 and a pip of 20 and you then adjust your peep pressure to 6 you will automaticaly have a pip of 22.

when you set the pressures up make sure you min alarm is all the way down and your max is all the way up and the cycle fail is above your peep but not your pip so you dont get the vent dumping thinking is an over pressure.

I think you are getting an high alarm what is causing the vent to dump the pip and thats why your getting that peep, not sure need more info

Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Dear charles & obiwan
Thanks a lot. Got it working. The PIP control pneumatic was faulty.
Met with SLE engineer in Arab health 2012, he helped us a lot.
Now only blender problem remaining, that is always whistle alarm even if both gases at same input pressure.
Thanks again for your replies.

Joined: Oct 2007
Posts: 75
Adept
Offline
Adept
Joined: Oct 2007
Posts: 75
sounds like a sticking shuttle but that may have caused your fluctuating pressures as it will vent cass through the alarm

Joined: Oct 2007
Posts: 75
Adept
Offline
Adept
Joined: Oct 2007
Posts: 75
cass = gass


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 11,691 guests, and 226 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,282
Posts74,585
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5