Hi,

Does anyone know how to check an spirometer Microlab 3300? I have two calibrate syringes (1Lt and 3lt).

Best regards,

Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
The manual says -

Quote:
To check the calibration of the MicroLab a large syringe should be used. This should have a 30mm outlet and be of at least three litres in capacity and preferably six litres. A 3 Litre Syringe is available from Micro Medical as Cat. No. SM2125. The MicroLab should be set up to record a FVC manoeuvre, connected to the syringe with the minimum of adapters. The syringe volume should be injected into the transducer evenly, without pausing.
If, after checking, the MicroLab appears to require re-calibration then please refer to your supplier for service.


Hope this helps

Lee


Don't forget "we've never had it so good".
Joined: Apr 2005
Posts: 266
Likes: 1
Mofe Offline OP
Master
OP Offline
Master
Joined: Apr 2005
Posts: 266
Likes: 1
Hi,

Thank“s for your help. Can you send an copy the service manual.

Best regards,

Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Sorry I've only got the user manual, which i have sent to your email.

Lee


Don't forget "we've never had it so good".
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Yet again ... see Frank. smile


If you don't inspect ... don't expect.
Joined: May 2010
Posts: 13
Novice
Offline
Novice
Joined: May 2010
Posts: 13
Hello

when you do check with the syringe should be borne in mind the following:

The measures taken have to be converted to BTPS conditions (Body Temperature and Pressure Saturated). This means that you can measure at least the room temperature and atmospheric pressure to calculate the BTPS factor.


Ever happened to me make calibrations without this conversion and then had the patient's values ​​wrong.

You have here a link to calculate the BTPS factor http://www.medicine.mcgill.ca/physio/vlab/resp/BTPS.htm.

All the best


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 1,753 guests, and 26 robots.
Key: Admin, Global Mod, Mod
Newest Members
j9_PLC, nece, Vitya, Shenzhen007, Eng. Craig
10,357 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,357
Most Online37,242
Apr 12th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5