Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Aug 2000
Posts: 300
Master
OP Offline
Master
Joined: Aug 2000
Posts: 300
We are presently using the APC Medical Model: 4170 Pacemaker. Just received a notification dated the 1st September 2011 from the USA parent company giving a recommendation that these devices should be replaced every ten years.

One of my staff contacted APC Medical UK for spare parts and was asked the age of the device. On stating the age to be eleven years a response of it should not be repaired but replaced was given.

I have contacted the MHRA on this at least to state that we all need to be notified of this 'recommendation' and make the appropriate decision around it. This will vary between one Trust and another no doubt.

Do you guys know of any other external temporary pacemaker devices on the market which may be similar to the Model: 4170? Remembering of course the need for comprehensive service documentation and access to all spare parts (PPQ question 5 - all six parts)

Thanks

Jim


Jim Gavin
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
It is not as if it is a mechanical device that "wears out". Is there any reason/justification from APC as to why it needs replacing or which bits fail? If you knew that you could do planned maintenance and replace those parts.
Robert
Or is it just a sales ploy.......?

In answer to your question, it is the first time I have heard of any electronic device "wearing out".


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Originally Posted By: RoJo
Or is it just a sales ploy.......?

Surely the word is "scam"? whistle

OK ... some electronic components may drift a bit with age (thousands of on-off, warming up, cooling down cycles) ... but that's one of the reasons that kit needs to be calibrated every once in a while, is it not?

@Jim:- are these things that complicated that they can't be repaired or cal-checked if need be "in-house", or by any half-decent biomed services company? think


If you don't inspect ... don't expect.
Joined: Aug 2001
Posts: 797
Likes: 1
KM Offline
Philosopher
Offline
Philosopher
Joined: Aug 2001
Posts: 797
Likes: 1
Hi Jim,
As you'd expect (clues in the name of the hospital) we've got loads of external pacing boxes.
Medtronic 5388 dual chamber, being the flavour that suits our taste best.

Joined: Aug 2000
Posts: 300
Master
OP Offline
Master
Joined: Aug 2000
Posts: 300
Hi Chaps

Thank you for the replies on this. The inference from APC Medical Ltd is with regards the passive components, namely the resistors going high and the capacitors drying out, leakage, potentiometer wear and tear, etc.

I obvioulsy agree with you chaps, carry out an annual or six monthly check as per the norm. Replace any faulty, defective or out of specification parts/devices. Probably the usual USA concept of litigation avoidance issues.

In reply to your question Geoff, they are most certainly not complicated, though again the USA contingent are perhaps of the opinion that they are. I have voiced my opinion on this quite robustly with explanations. I have been maintaining this manuafacture of external pacemaker since 1975 (37 years), simple little criters that they are.

As stated previoulsy I have raised this to the MHRA as an adverse incident report, because we at least need to be aware of this matter and deal with it accordingly, as we each see fit.

Karl, I will look at the Medtronic range as we only require single chamber pacing. I will pick your brains on relaibilty, service documentation, spares, etc at our managers meeting in Halton General Hospital on the 17th April, I'm bringing the biscuits I believe.


Reagards to you all.

Jim


Jim Gavin
Joined: Mar 2001
Posts: 40
Technologist
Offline
Technologist
Joined: Mar 2001
Posts: 40
Hope you get somewhere with the MHRA Jim. In my experience they default to supporting the manufacturers position most of the time. If so then it seems all critical support equipment should be binned on it 10th birthday!


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 8,703 guests, and 27 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,482
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5