Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 3 1 2 3
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Following on from an earlier thread ...

It seems that (in the USA) the Centers for Medicare and Medicaid Services (or CMS, for short) has issued an edict (.pdf) which seems to require that (in order to receive Medicare and Medicaid payments) hospitals need to adhere strictly to manufacturer recommendations for inspecting and maintaining critical care and imaging equipment.

This statement has apparently drawn considerable comment and concern in the States, as it contradicts the more flexible stance recently taken by the Joint Commission* (which has allowed hospitals to use their experience to determine how often medical devices should be inspected). This flexibility has been widely supported in the literature and by biomeds in general (and, I dare say, by certain members of this forum)!

Meanwhile both ASHE and AAMI have been drawn into the debate (as indeed has ECRI) ... so hopefully we shall at last be seeing some "final" words of wisdom on this issue (in the US, at least).

But as it stands, the Bottom Line of the CMS requirement is:- PM must follow the manufacturer’s schedule on critical equipment, and some form of evidence-based scheduling on non-critical devices.

To my mind, that sounds like good policy ... especially the second part. smile

* Joint Commission on Accreditation of Healthcare Organizations


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Just in case anyone was wondering:-

AAMI
ASHE
ECRI
JCAHO


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Looking again at what CMS is saying, they make it very clear that even where revised scheduling (of PM) is permitted, it the interval - not the procedure itself - which may be adjusted.

That is (and I quote):- "alternative equipment maintenance methods are not permitted".

And that:- "hospitals must continue to follow the manufacturer's recommended techniques for maintaining equipment, even if the hospitals alter the frequency of maintenance activities".

But OK ... PM intervals for "non-critical" [definition required] equipment may be varied (lengthened, presumably), just as long as such a change is supported (justified) by "evidence".

What form does this "evidence" take I wonder? think

That is:- how do we satisfy the Managers, the Bean Counters, Lawyers and Miscellaneous Suits ... et al (and anyone else who routinely stands in the way of progress in general)?

Meanwhile, here is some more on EBM (Evidence-Based Maintenance) (.pdf). smile

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

This is what CMS says:-

Quote:
If a hospital is adjusting non-critical equipment maintenance frequencies to be below those recommended by the manufacturer, such adjustments must be based upon a systematic, evidence-based assessment. The hospital must document the assessment for all equipment with less frequent maintenance than what the manufacturer recommends, including the selected maintenance stategy that led to the lower adjusted frequency, and supporting evidence. The evidence must provide support that the frequency adjustment will not adversely affect patient or staff health and safety. The assessment of whether it is appropriate to use a maintenance strategy that results in less frequent maintenance than what the manufacturer recommends must be performed by qualified personnel. In the case of medical equipment, a clinical or biomedical technician or engineer should be considered qualified.

Straight away I can see that we need to agree on what is meant by "adversely affect"! think

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
Why employ intellegent, knowledgeable Clinical Engineers when you can just have a monkey to follow the instructions given by the manufacturer? Or even put it on contract and save a wage?

And if the manufacturer says you have to replace a non-worn out part which you can only get from the OEM every time you look at it, you have to because the powers that be say so.........Money making on a plate!!!!!!!

Robert the cynic


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Furthermore:-

1) Manufacturer's service recommendations are written well before equipment goes into service. In other words, they are simply an informed (hopefully) guess (and rarely revised following years of in-service experience)!

2) How many times have we come across equipment of similar type, but from different manufacturers ... only to find service recommendations for one being well over the top, the other "hard to find"?

3) Surely it is clear that the medical equipment scene is driven by the manufacturers ... and not the biomeds ... and that the "Regulatory Bodies" are clearly on the side of the manufacturers (aka, CYA and taking the easy line of least resistance). frown

4) This sort of thing is yet another reason why biomeds need to be ... er, assertive and keep fighting the gougers! whistle

How can either the manufacturer (or, for that matter, the Regulator) know "what is best" at any individual hospital or clinic? At best, the manufacturer's recommendations should be viewed as simply a basis (or guide) to locally developed procedures (whether they be "less", or "more").

So ... beat them at their own game, and present the evidence for Evidence-Based PM! smile

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

And how would I go about that? For starters (and as I have said before) I would propose (as candidates for lengthening PM intervals) all equipment that didn't need a repair [definition required] at all of the previous three (say) PM visits. smile

Joined: Aug 2007
Posts: 53
Scholar
Offline
Scholar
Joined: Aug 2007
Posts: 53
Geoff ,
I can only speak about the manufacturers I have worked for and the service side of their business was not deemed a profit generating entity.

My experience has been is that there are small number people ( in companies and hospitals ) who , at many levels , manage to sell the “ what if this goes wrong “ or “ this could break if …!” and before you know it there is a policy or a standard , most of us are powerless to address it ( largely because biomeds in general are a fragmented group) and then we all stuck with it .

Biomeds can be their own worst enemy, the cost of a Biomed Department in hospital is not big expense – relative to the cost of running a ICU Bed or a ED resus cubicle it very small but we all continue to tackle the issue of cost as our contribution to helping the hospital, now I’m not saying don’t save money but that’s not the focus of the clinical staff when they treating a patient.

That link you gave us to the Binseng PPT highlights this [some more on EBM (Evidence-Based Maintenance http://ohiocea.files.wordpress.com/2011/03/wang-huffman-evidence-based-maintenance.pdf ] ( in my opinion ) – if you look at slides 42 to 45 ( Defibs ;PCAs;Vital Signs ;Blanket Warmers ) - NPF – No Problem Found was by far the most pronounced issue there – and I would think this would be the issue of discussion more so than any of the other issues , however that point seems to be trivialized – but that’s the issue which cost the hospital the most in direct costs ( clinical staff time ) and when things go wrong there is the human cost ( the issue closest to the clinicians heart hence using a $3000 catheter as single use item doesn’t bat an eye lid )

I have posted a suggestion on the NAMDET site [ http://www.namdet.org/da/115909 ( Intro The Clinical Equip……..) ] but if biomeds , who are best positioned to address the NPF issue , embraced this as a issue via the system ( using safety & quality and consistently measuring NPF as a reportable metric ) then how often you do or have to service a device almost becomes a non issue .

The performance of equipment fails as a result of incorrect use far more than it does from a engineering / technical perspective – and it’s not a dig at anyone, we all make mistakes – it’s a patients safety thing , the technology gap / issue is getting larger .

The real issue is that clinical staff are seldom told of these “user issues” [they not consistently report and measured ] and so users tend to do the same thing over and over ( and hence the same issue come up all the time) . Even worse is when high profile clinicians are involved, not only can they cause the most damage as they tend to do the most difficult cases so if ever something was going to go wrong , that’s when it does . And we all know the reporting of these incidences is dismal.
It’s a bit long – sorry

Enjoy the weekend , Cheers George

Last edited by GeorgeK; 18/05/12 6:16 AM.
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Weekend? Did I sleep too long last night (or better still, wake up in the Magic Kingdom)? I wish! whistle

Thanks for that input, George (and long is OK ... by me, at least). And surely all that effort deserves a plug (it's a nice approach)! smile

I'll quickly take your paragraphs one-by-one, if I may:-

1) Perhaps. But there have been many examples of gouging when it comes to selling parts!

2) Indeed. Sadly, we have more than our fair share of Jobsworths here in the UK.

3) Yes. Biomeds are generally Good Folk ... who often "try too hard" (to please, and all the rest). There has always been far too much focus on cost and/or "saving money". We should be looking at other ways of measuring and presenting the value of biomed services (as has been said on here over, and over again)! In short, biomed should be presented as a Quality Service (that is, one that meets the needs - in technical support terms - of the clinical users).

4) Yes. Agreed. "Finger trouble" = more training required! But at least the users there have (presumably) reported (what they believed to be) faults in order to have all those Work Orders raised:- I see that as a Big Positive.

5) Nice one. I recommend everyone to download that .pdf (and critique it if need be)! Meanwhile, can I précis it down here to that well known cry:- RTFM! think

6) Yes. In my experience it has always been like that. It's always be easier to simply "blame the kit", and the present day "No Blame Culture" hardly helps. You know, "Lessons Shall Be Learned" and all that [censored]! On the other hand, safeguarding patient safety is what Biomed is all about.

7) As I may have mentioned before, sometimes (often?) biomeds need to be, shall we say, assertive - and never be afraid to Take the Bull by the Horns as and when required!

But what about Evidence-Based (and its close associate Risk-Based) PM, George ... do you go for that?

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
OP Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Originally Posted By: GeorgeK
... but if biomeds , who are best positioned to address the NPF issue , embraced this as a issue via the system ( using safety & quality and consistently measuring NPF as a reportable metric ) then how often you do or have to service a device almost becomes a non issue.

Sorry George, but I don't see what your're saying there. think

PM intervals cannot become a "non-issue" if they are being mandated by Regulatory Bodies!

Meanwhile, NPF is often just another "fault code" in my experience (and therefore already "reportable" - potentially, at least). Such codes are a response to Work Orders or RFS (Requests for Service), surely (and not PM)?

I may be a bit dim (more tea required), but I can't see how that impacts on the PM interval - especially (as I keep banging on about) if the "Risk-Based Model" is applied.

As I have already said:-

Originally Posted By: Geoff Hannis
I would propose (as candidates for lengthening PM intervals) all equipment that didn't need a repair [definition required] at all of the previous three (say) PM visits.

But this applies to PM visits not Work Orders. And if it were to be applied to PM results, then an accumulation of NPF codes should serve to trigger an extension of the PM interval (in a Risk-Based, auto-adjusting system). In passing, note that it also works in reverse:- PM intervals could be shortened if so many consecutive PM's did require a repair (intervention, whatever is decided). What we are seeking is PM equilibrium - or if you like, the "Goldilocks Condition"! smile

Page 1 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 2,099 guests, and 399 robots.
Key: Admin, Global Mod, Mod
Newest Members
Mark B, Royal, davidarhui, Czr, Jimmy Li
10,391 Registered Users
Forum Statistics
Forums26
Topics11,278
Posts74,590
Members10,391
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5