Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Dec 2002
Posts: 20
Simpo Offline OP
Dreamer
OP Offline
Dreamer
Joined: Dec 2002
Posts: 20
Gents

Can anyone throw some light on the method for carrying out the dry side calibration test on the Omnifuse syringe driver using the Omnifuse software and a Fluke IDA 4 Channel tester. I'm being asked to input an ambiguous measured value in KG?? and only have the standard IDA options of mmHg, cmH2O etc.

Any help appreciated

Paul

Joined: Oct 2003
Posts: 51
Scholar
Offline
Scholar
Joined: Oct 2003
Posts: 51
Hate to be a party pooper but the short answer is you can not.

To complete the dry side occulsion pressure measurement you need to use the the approved pressurement tool, - a load cell (mounted to a dummy syringe) (dry side occulsion pressure is a direct measurement of the force being applied by the pump drive system to the syringe plunger) with the relevent control box that allows you to zero the load cell etc. This system then gives you the applied force values to enter into the omnifuse calibration program to allow the right values to be written back to the pump software.

The tool is available from Smiths but it is made by Force Measurement Systems in Glasgow, who may be able to help you.

Joined: Aug 2007
Posts: 306
Master
Offline
Master
Joined: Aug 2007
Posts: 306
The part No. is 0151-0287 load assembly kit.
For the size gauge set if you are calibrating 0151-0259 (sizes 8-35mm)

You will also need the software depends on your software issue in your pumps, contact the techs at smiths for the one you will need.

Joined: Feb 2004
Posts: 14,813
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,813
Likes: 72

Joined: Dec 2002
Posts: 20
Simpo Offline OP
Dreamer
OP Offline
Dreamer
Joined: Dec 2002
Posts: 20
Many thanks chaps, lets see if Smiths will sell the test equipment now!


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 3,321 guests, and 14 robots.
Key: Admin, Global Mod, Mod
Newest Members
Danny1621, zeedox, Eman Abdallah, OmarLY, Hans Pauwels
10,384 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,576
Members10,384
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5