Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 4 1 2 3 4
Joined: May 2001
Posts: 193
Chris-H Offline OP
Mentor
OP Offline
Mentor
Joined: May 2001
Posts: 193
Following a reported fault of electrical failure alarms in theatre I was summoned to inspect a Sony video monitor model PVM1443MD.

The mains lead plug, wiring, and controls all were intact and nothing untoward.
Being a class one device, tested the unit with a Metron Electrical safety analyser to IEC 60601.1. test regulations.

The unit was additionally tested using a Seward PAT1000s electrical portable appliance tester.

Never did it under any circumstance display a fault with either of the two mentioned safety devices.

However upon return to theatres, plugging this monitor back into the Isolated Power Supply in Theatres, the ITPS monitor (IPSM002-MB) displays a killo ohm fault with constant alarm.
This then lacthes a fault conditon that has to be reset at the main power intake board (Consumer unit).

The device instaled at the intake (Consumer unit) is an IPSM002-MB.
What this device does is monitor the insulation resistance of the unearthed AC network, mainly found in group 2 medical locations.
The current temperature of the isolation transformer is permanently measured. It is in the form of a DIN rail mounted device.
Its function is to measure DC voltage that is superimposed on the network. Any Ohmic insulation fault path between the network and earth is detected.
If the reading is below the selected response value the relay K1 and the Kohm LED will operate. The bar graph indicatior permantly shows the system insulation level to earth in Kilo Ohms.
The current passing through the isoaltion tranformer is measured via an external CT. If the current exceeds the adjustable reference value, the relay K1 and the alarm LED operate.
The bar graph indicator permantly shows the percentage in use.

After consulting the manufacturer (Starkstorm) they are of the opinion that the SONY Video monitor I had plugged into the Theatres mains wall outlet is still at fault.

How can this be I argued if the monitor passed both IEC60601 and PAT testing?? confused

Anyone here have similar IT power systems in their theatres giving spurious or odd results?

I'd be inclined to alter the reference value on the (IPSM002-MB) however Starkstorm do not recomend this as they state their intallation currently meets with IEC regs governing installation in Group 2 medical locations, and altering this value would void this regulation / installation ?? eek

Bloomin suss that one eh ? mad

Chris-H


Ours is not to reason why?,
Simply obey & then comply !
Joined: Apr 2001
Posts: 266
Likes: 5
Master
Offline
Master
Joined: Apr 2001
Posts: 266
Likes: 5
At an absolute stab in the dark there could be a remote possibility of....

If you are using the BNC connected Shielded connectors on the monitor then maybe a few loose strands from one of the video signal leads are finding their way onto the outer earth shielding of the cable and back through the chasis earth.

Might be an idea to see if it alarms with the cables removed so you can isolate the device from the interconnecting leads.

Just a stab in the dark I know, but its worth a try if you can think of nothing else, these nurses have a tendancy to run over cables and generally abuse them.

Joined: Aug 2000
Posts: 300
Master
Offline
Master
Joined: Aug 2000
Posts: 300
Chris

Another stab in the dark so to speak, does a trip occur in the mains monitoring device immediately after the Sony monitor is plugged in?

Based on a reply of yes, rather than later, have you looked at the mains rf/emc filter in the Sony monitor? Our theatres are not mains monitored and as such no tripping problems using Sony monitors.

Could be the mains filter at switch on, surge of the X/Y, particularly the Y capicitor in the filter?

Jim


Jim Gavin
Joined: Oct 2000
Posts: 36
Visionary
Offline
Visionary
Joined: Oct 2000
Posts: 36
Chris -

Can I ask a few questions ....

If I understand correctly, of the the three IT monitors - isolation, overload and transformer temp - it is the first of these that is triggering the alarm.
What is the kohm refernce value at which the isolation alarm is set ?
Can the alarm be cleared without unplugging the offending monitor ?
What measurements were included in yuor Metron/PAT-testing of the monitor ?

Thanks,
Ged Dean

Joined: May 2001
Posts: 193
Chris-H Offline OP
Mentor
OP Offline
Mentor
Joined: May 2001
Posts: 193
Soo many questions and so little time/knowledge confused
1) Yes as soon as the monitor is powered up, whether it be from the switchable mains socket or either the power button on the front of the Sony monitor, results in it tripping. It is indescriminate.
2)Have been unable to try the unit as of yet with all BNC's dissconnected. However the BNC at rear of monitor was used with refence to exposed parts during electrical safety testing and passed with flying colours.
3) The Kilo-Ohm reference on the IPSM002-MB (Starkstorm bit of main supply monitoring) has a "resistance to earth setting" of 240 kilo-Ohm. with a "current setting" of 20 amps. This is what has been stated in the commisioning report.
4) Yes I agree, it could well point to an internal mains filter, however, surely this would have shown up in the Metron Electricval safety test, this was carried out with earth leakage current limit of 1000 micro amp(open supply and reverse mains) 500 micro amp(normally closed and norm closed reverse mains), enclosure limits of 500 micro amps and (100 normally closed and reverse mains) as per Default setting by Metron for class one devices IEC60601.1.
So I think not ! eek
4)No the alarm cannot be cleared without unplugging the offending monitor. It is a latchable alarm that has to reset at the main intake (Estates sparkie).
5) Unfortunately I do not have the PAT results but I vaguely recall a limit of 0.5 ohm being entered ( As do not regularly use this bit of kit)

Chris-H


Ours is not to reason why?,
Simply obey & then comply !
Joined: Mar 2002
Posts: 117
Savant
Offline
Savant
Joined: Mar 2002
Posts: 117
Chris, your isolation monitor sounds very comprehensive in what it measures. In this situation though, only the impedance measurement part is up for consideration of course. From the information you have given us, this is very definitely an equipment-derived (mains) line-to-earth impedance problem, unless you already have an isolation monitor/system problem (see testing method below).

If you have not already done it, even though such a low impedance (1K-ohms) is being registered when the video monitor is connected, suggesting the problem is almost certain to be fairly and squarely in the video monitor itself, I would suggest you do a quick check of the system integrity with respect to the overall ‘leakage’ of the system. Isolation systems can deteriorate unnoticed and should be checked on a regular basis anyway. This will give you to some extent ‘peace of mind’ that your isolation monitor is calibrated and doing its job properly, or not, as the case may be.

If you do not have an isolation system checker, then build yourself this quick piece of test kit I have outlined below. With this you can test for that 240K-ohms alarm-trip point you may still have your eye on adjusting! You can also measure the isolation leakage (current flowing) from both lines with respect to earth. (I will refer to ‘Lines’ now, not ‘live and neutral’ in the following explanation)

First off though, disconnect ALL the equipment in the theatre suite in question, from All the sockets in the wall/ceiling mounts etc. (Don’t forget wall-mounted viewing boxes, surgical angle lamps etc. etc. if plugged into the ordinary sockets i.e not hard-wired in). With this done, the isolation monitor should be reading a high impedance, say at least 4-500K-ohms? if the ‘trip-point’ of your isolation monitor is 240K. If the reading is still much lower, say around the 240 value, then you should attempt to find out why, as it might not take much to trip the alarm when extra equipment is plugged in later. Generally, isolation monitors are designed to alert the staff with a current limit of around 2 to 5mA. At your mains voltage and for this trip impedance (240k), this would equate to an alarm tripping point of just 1mA. (If it had not been showing as low as 1k-ohm when you plugged in your video monitor, I would have suggested you look for an accumulation of impedances that might have been just ‘tipping the balance’ over the 240K trip limit. But this reading is LOW!).

With the isolation monitor reading hopefully well above the 240K value and with all the sockets empty, I am assuming that if you plug in your video monitor now it will show approximately the same 1k reading and alarm - Yes?? If the reading stays low, and then returns very high when it is unplugged again before the alarm is reset, consider the video monitor as a very likely 'culprit’. Unplug the monitor again.

Now to test the system. Get a 240 volt light bulb connected to two wires and momentarily touch them across the two line (live & neutral) conductors in any mains wall socket to ensure it lights up. Then connect the two wires, one to a line conductor and the other to earth, in the same socket. It should not light up. Connect to the other line conductor and earth. In neither case should it light up. (If one line to earth does light, you do not have an isolated system any more i.e one side is already down to, or very near, earth potential. If this is so, stop here immediately and investigate the reason).

Note: The isolation monitor may alarm each time the bulb is connected from each line to earth, just reset and continue. Do not forget this light bulb check step.

If the light-bulb test gives the correct results, then next get an ammeter (it will be measuring only a low ac value) and measure the isolation leakage current (impedance-equivalent) at the same wall outlet, of each line to earth in turn. As you did with the light bulb, connect it between one line and earth (not between line and line in this test, of course!!) note the reading and then do the same between the other line and earth. Note that reading too. Now calculate the individual impedances from the individual current values of each line to earth. The highest current value (lowest impedance) reading should equate to the isolation monitor’s ‘estimated’ and displayed value of isolation impedance in the system. Note: for trouble-shooting purposes, the leakage current value you read in either line when doing this test, relates to the isolation value of the other line-to-earth and not the line-to-earth that you are measuring.

Now with the trip-point potentiometer tester you have built (see below), plug it into the same wall socket (or any other). Slowly reduce the pot resistance until the isolation monitor alarms on the line selected. When it alarms, note the impedance reading on the isolation monitor and then remove the mains plug and measure and note the resistance between the relative power pin on the plug and the earth pin on the plug. Now return the pot to maximum resistance and switch over to the other line. Plug it in again and do the same measurement and note the same two readings. Using these pot resistance readings you can firstly, with the help of dear Ohm’s law (and 240volts), simply work out the current that has tripped the alarm. It should of course be near the 1mA level in both cases. And then secondly, check that the impedance readings found on the isolation monitor compare favourably with the resistance readings on the pot. Assuming that the leakage current readings, you took above, were near enough the same, it should be approximately the same reading i.e trip the alarm around the same point on the pot. This will give you an indication of how accurate your isolation monitor is (in the impedance measuring mode only) and hence if you have a problem actually in the isolation monitor/system itself. The pot readings should be around the 240K-ohm mark if the isolation monitor is working correctly. (If not, it might also help to decide whether the pot you wanted to adjust really needs adjusting, though I would still recommend consulting the company first).

Now to return to the video monitor itself. As you said, you did the insulation impedance of both live and neutral together, to earth, on the video monitor and it passed OK. However, the main point to note here is that the Insulation Test is usually done with a DC voltage in testers, which might well leave undetected (as has already been suggested) say a ‘leaky’ interference-rejection capacitor (normally having mains-frequency voltage applied across it) in what in your case here (if the video monitor is not, like me, of the ‘old generation’) could easily be a switch-mode PSU.

I was going to suggest that you had possibly over-looked, due to perhaps limitations of your testers, doing the earth leakage test in the mains reverse connection. The reason being (for interest only) that in the normal mains configuration testing on a non-isolated supply, any caps etc. going to earth on the neutral side that might be leaky, would definitely not be detected due to lack of ‘electrical stress’ (normally produced by the live wire to earth) needed to produce a current and the earth-neutral link. But since you have done that particularly important test in this case, the two things that might be happening are firstly, that when you switch the video monitor on, the initial charge-up surge current (though very small) into the mains input filter caps (tied to earth) is causing the isolation monitor to momentarily trip but permanently latch the alarm - a condition the above system tests would be unable to reveal of course. (If the reading stays latched at 1k after straight away removing the video monitor from the mains socket, this might help to confirm the theory). If, however, the impedance reading does change back to high impedance when removing the video monitor whilst in the alarm condition, it might strongly suggest that you have a ‘hard’ fault to earth not a transient impedance dip. In this case it should be easier to find. The second possibility, and perhaps the more likely, (if it is an SM-PSU) is that one (or both?) of the MOV’s (metal oxide varistors) may be too sensitive (or too low a value) and momentarily triggering low impedance to earth at switch-on.

Now to the construction of the isolation monitor leakage alarm-trip tester:
Get yourself a linear (not log) potentiometer of say 470 kilo-ohms and a center-off, single-pole, double-throw mains switch. Connect a short piece of any size (a one or two-amp rating will do) three-core mains cable to the live, neutral and earth of a 13-amp plug (any rating fuse in it will do, as it will not be there for protection, only continuity) - or whatever type you use in theatres. Connect the slider of the pot to one of the other contacts on the pot as well as connecting a wire from this jumpered pair first to one end of an in-line fuse holder (with say a 100 or so mA fuse) and the other end of the fuse holder to the earth wire of the cable. Set the pot to read maximum ohms with respect to the other free terminal on the pot. Now connect this free terminal on the pot to the center pole on the switch. Connect now one power line wire in the cable to one of the two spare switch poles and the other power line in the cable to the other spare switch pole. Tape over any bare wires.
Note : Even though it is generally safer working on isolation systems, in certain circumstances (particularly across the two line conductors) it can still be dangerous, so use the usual amount of safety awareness when doing these tests.

Lastly for my own interest, did I interpret you correctly when you say that the isolation monitor is looking to measure dc current superimposed on the (ac) power circuit connected to the isolation transformer? Or did you in fact mean that it uses a superimposed dc to do the impedance measurements from the lines to earth ? The latter (if pulsed) to me would probably be more likely as it would need some form of a ‘changing’ current to measure not only purely resistive components to earth, but capacitive as well. If I am wrong, what dc is it measuring on the system? Measuring the temperature of the transformer in an isolation set-up was new to me as well. Interesting. Seems a very thorough monitoring device.

Finally, permit a note from an ‘old-hand’ on pot-adjusting. If the pot in question that you thought about adjusting is not easy to get at i.e. need to remove covers, it was good initial advice from the company. In fact as a basic technician tenet, it is never a good idea to adjust internal ‘pots’ in any piece of equipment as a way around problems, however tempting it may be, unless you have explored all the other avenues to try to correct the problem first. Although ‘pots’ can go noisy and upset readings (or be mis-adjusted by others), best wait, as one can easily end up with more problems than bargained for particularly in this case if you have to have the isolation monitor re-calibrated with specialized equipment later (at the hospital’s own expense?) having found the fault elsewhere.

If now you decide it is in the video monitor, (don’t rule out the small possibility that the isolation monitor may be reacting incorrectly to spurious switch-on dips), concentrate only on the PSU, (old style or smps) . If it is a switch mode, I would suggest (only to test of course) just removing the interference caps that go to earth and then MOV’s or vice versa.

Keep going.

Joined: Mar 2001
Posts: 208
Ken Offline
Master
Offline
Master
Joined: Mar 2001
Posts: 208
We have recently started using new theatres with isolated power supplies and are starting to have problems with the mains triping out.
Can this problem be caused by plugging some equipment into none isolated power outlets.
Where can I get more information about the use of these isolated outputs.

Joined: Jul 2002
Posts: 123
Savant
Offline
Savant
Joined: Jul 2002
Posts: 123
Chris/Ken

Chris, historically we have had to fit isolation transformers to endoscopy trolleys due to high leakage currents associated with Sony monitors. Would appreciate you highlighting the eventual solution to your problem.

Attached are a couple of references that may help illuminate this topic. Looks as if we will all require to gain more understanding in this area.

www.brandon-medical.com/nletter/issue4/safe_electrics.htm

IEC INTERNATIONAL STANDARD 60364-7-710 First edition
Electrical installations of buildings – Part 7-710:
Requirements for special installations or locations – Medical locations
Reference number
CEI/IEC 60364-7-710:2002

Bill


Bill
Joined: Oct 2000
Posts: 36
Visionary
Offline
Visionary
Joined: Oct 2000
Posts: 36
I think that what is normally being monitored in IT-systems is in fact insulation, rather than isolation. (I’m told that in some parts of the world the practice is to measure isolation impedance rather than insulation resistance, but then the measurement is normally given as a prospective mA current value rather than a kohm value).

If your insulation resistance is being measured by a dc monitoring signal then any errant dc present (eg: from a faulty s-m power supply) will add or subtract to the measured value, causing over- or under-sensitivity respectively. In the former case this could result in nuisance alarming, the latter in loss of safety margin.

Potentially relevant to your case is that in the Bender sales literature, (Bender are also manufacturers of IT-power systems), they report having found a particular problem with a number of makes of VDU appliances. Apparently a certain internal EMC fix - of which unfortunately I have no knowledge - used in some of these devices results in small dc current flow of the type described above, which then affects results of the IT insulation monitoring as described. If your Starkstrom IT-system uses a plain dc monitoring signal then this might be the source of your problem. (I believe Bender now use a dc signal which is pulsed at very low frequency, which then enables the monitor to detect the presence of any other dc signals and correct results accordingly).

p.s.: With regard to comments on IT-system monitoring above , insulation, overcurrent and temperature are all requirements of IEC-60364-7-710 so will now be the norm. (The relevant chapter in the IEE’s Guidance Note 7 (2003) supporting the ‘Wiring Regs’ is based upon this IEC Standard).

Joined: Mar 2001
Posts: 208
Ken Offline
Master
Offline
Master
Joined: Mar 2001
Posts: 208
We are trialing a new camera system and associated components that is wired through an isolating transformer.It is since we began to use this equipment that our problems started. Unfortunately the system does not immediately trip when plugging the camera system into the mains supply, but this is the only common piece of equipment being used when the problem occurs.
All leakage tests on the camera system have given the expected results. It could be that the monitoring system is faulty and I am awaiting for this to be tested.

Page 1 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 2,869 guests, and 13 robots.
Key: Admin, Global Mod, Mod
Newest Members
j9_PLC, nece, Vitya, Shenzhen007, Eng. Craig
10,357 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,357
Most Online37,242
Apr 12th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5