Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
#63608 31/01/13 10:58 AM
Joined: Mar 2002
Posts: 188
Ed SWM Offline OP
Mentor
OP Offline
Mentor
Joined: Mar 2002
Posts: 188
Hi All,
Just wondering if anyone out there has had any trouble with recent batches of admin sets for the Asena GW pumps. We have had a number reported as coming up with "BAD SET". Further investigation showed the pump to be within spec but after requesting an example set, we found the set compliance to be "feelably" different the newer sets being slightly firmer.
I have spoken to carfeusion but it would seem we are the only people with this problem!
Has anyone else had any experience of this problem or do we have a rogue batch or batches of sets?
Regards
Ed Bennett
Southwest Medical Ltd

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Although the newer set may well be a bit firmer, I doubt that the "real" (underlying) problem there is the set, Ed. I believe it will be more on the "pressure sensing" side of things, on the pumps themselves.

Unfortunately, I never got around to cracking this one (see my post 07-May-11).

But (in short) there has to be a way of adjusting the "BAD SET" sensitivity, if you see what I'm saying. You need an "address" - but my own discussions with Basingstoke came up blank on that one (I was after a complete list of the address codes - but I seem to recall them saying something along the lines that "only Germany has them"). whistle

But I suspect that "they" don't want folk being able to adjust their GW's to render them less sensitive to set compliancy (and therefore opening up the possibility of being able to use "generic" sets), in order to protect sales of "original" (genuine, whatever) giving sets!

I had better not say any more about my "experiments" with the GW at around that time, lest some "wag" comes on here and accuses me of "hacking", or some such thing. frown

Meanwhile, you say that the pump is "in spec" ... but which spec do you mean? You might get away with having another go at the pressure sensor cal (Code 17).* Sometimes a bit of tweaking is required, especially with older (worn, tired perhaps) GW's.

* And you really need a Druck or some such thing to make a good job of that.

Joined: Jul 2010
Posts: 6
Newbie
Offline
Newbie
Joined: Jul 2010
Posts: 6
Hi Ed,

Just to let you know, you are not the only one to have had this problem. We contacted Carefusion about these new sets in mid December (I believe we are only one of a couple of Hospitals in the UK still using GW’s). We also encountered a similar problem (alarming ‘Hi Pres’ error 4, rather than ‘Bad Set’ error 6) the week before Xmas with lots of our library GW's coming up with pressure faults. It was only after talking to the staff on the wards about the sudden influx of these that we found out that Carefusion had changed the "giving sets" without telling anyone. When we flagged this up to the company that we were having a big problem, their reasoning for the change was due to the original 'softer' sets containing some carcinogenic compound in the plasticiser, taking this out making the new sets harder.
We got one of their principal engineers to come in to check the problems we were having, after confirming there was indeed a problem, the solution was to alter the ‘Pressure setting ” from Normal to High, some also required the “null pressure setting” to be reset (11 +/- 4) and to recalibrate the pressure sensor. This seems to have alleviated most of our problems, but their engineer said there was a “fix” that CareFusion had come up with that involved the replacement of the plastic bushes in the door hinge. We haven’t tried this as yet so don’t know if that will solve the problem. Hope this is of some help smile


Mark

"Its nice to be important, but its more important to be nice"
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

That's an interesting post, Mark. smile

Do you record (keep a note of) those "null pressure setting" pot R2 (12.8) adjustments? My own notes suggest quite a variation (but I only have notes from four pumps). I would be interested in your data!

But isn't it "great" when manufacturers just blunder on and change a major component of a system (because that's what the giving set is, in this case especially) without telling (warning) anyone about it? whistle

I can't help wondering if they actually tested a few of the "improved" sets (on real GW pumps), as well!

Although there may not be many GW's in UK hospitals, there are quite a few around elsewhere (at vets, for example).

As discussed before, the GW's are always a bit fussy about "pressures". But if Alaris (whomever) were a bit more "open source" about those code addresses, these pumps could be adjusted to run on more or less any set! That is, just by tweaking the appropriate Codes (addresses).

Lastly ... best tell the guys to forget about the door bushes. That's a definite Red Herring. OK, if they were totally shot, then the pump wouldn't hold its pressure - but that's not the "answer" here (which is, as I keep saying, to de-sensitize the "BAD SET" sensing - open up [widen] the spec, whatever you want to call it).

I have not seen (felt?) the "new" sets. Are they significantly harder (stiffer)? If so, that would be a pity, as it shall probably mean more GW's retiring earlier than they otherwise would have (due to being worn out, that is). frown

Does the new set have a new Part Number? My notes mention that "GW 273-005" was the set we liked to use.

Meanwhile, has anyone yet fathomed out how the "BAD SET" sensing feature actually works? think

I believe that the original idea of the "BAD SET" error message was to alert the user - and stop the pump - when the set had been loaded incorrectly (that is, back to front)!

Again, see my post of 07-May-11.

In passing, those who are interested in this sort of thing (everyone, surely?) might like to compare (the design, performance, circuitry, means of adjustment, and sheer versatility of) the GW with it's clone(s) ... the CME Nikki et al. Consider also if they (not only look very similar, but also) use the same motor-gearbox? Then ask yourself the obvious question:- which one came first? think

Joined: Jul 2010
Posts: 6
Newbie
Offline
Newbie
Joined: Jul 2010
Posts: 6
Hi Geoff,

We don’t tend to keep a record of the null pressure settings, but from memory, we have had some as high as 20, most seem to be around 15 or 16 if they are out, I tend to set it to about flicking between 10 or 11 as it seems to drift upwards a little. Depends upon the age / when we last calibrated them & if the pressure transducers have been replaced previously (age range of ours are from a couple of years to several).

We did receive a “Customer Letter” detailing the changes & new part numbers after contacting the reps about the problem in the first.
The part numbers are the same but with a V on the end (ie GW 273-003V)

Quote: “We are pleased to announce that the following of our GW-Infusion sets will be available in a DEHP-free version.
What is DEHP?
DEHP Di(2-ethylhexyl)phtalate is a well-established plasticizer for most PVC containing medical devices. CareFusion is continuously introducing DEHP-free products to reduce the risks involved with the use of DEHP”.


We also wondered if they actually tested these "improved" sets on real GW’s, as well!

We didn’t take the ‘door bushes’ “solution” too seriously!
(I am not advocating it at all, but you could always turn the “Set Test” feature off, may be for VET use. 

Yes, the "new" sets are slightly harder (stiffer) with a slight blue tinge to the plastic, more noticeable when you have an old one beside it.

I may have got the wrong idea but I thought the “BAD SET" sensing feature worked by the pressure sensor, sensing the backpressure created by the giving sets with the downstream blue filter “thingy” in them, as the old sets without these throw up a "BAD SET" error message. We have to remember, every time we change the set we use for pressure calibration, to chop out the filter thingy otherwise it doesn’t cal properly, can catch you out sometimes until you realise it hasn’t been done!

I agree I also think that the original idea of the "BAD SET" error message was to alert the user - and stop the pump - when the set had been loaded incorrectly.


Mark

"Its nice to be important, but its more important to be nice"
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Ah yes, the "blue filter thingy" (anti-siphon valve) ... yet another reason to disable the "BAD SET" sensing (in Veterinary use, of course).

Again, with all these "variables" available in the various (genuine Alaris) sets, and the seemingly "demanding" nature of the sensing technique, I still reckon the manufacturer should look at revising the manual, as it were. Open up the sensing spec, and/or make it easy to "dial in" the pump to suit the set being used (without having to re-calibrate). But, better still:- publish the darned codes! whistle

Does anyone know if the GW is still being manufactured (in Germany?)?

Lastly, what software version are we up to these days? And does that have any effect on the issues being discussed here? think

Meanwhile, I'm not too clear on that &#61516 code, Mark. Would you care to spell it out for me? You can always use a Private Message if you'd rather. smile

Joined: Jul 2010
Posts: 6
Newbie
Offline
Newbie
Joined: Jul 2010
Posts: 6
I'm not too clear either on the &#61516 code i think it was supposed to be a ')' end bracket not sure what happened!

As far as i know the GW's are still manufactured in Basingstoke UK, mainly for the overseas markets.

Software version that has just come out is V52A don't know what the difference is over the current version we have that is V51F


Mark

"Its nice to be important, but its more important to be nice"
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Originally Posted By: Mark L. Peacock

I'm not too clear either on the &#61516 code i think it was supposed to be a ')' end bracket not sure what happened!


In my haste I had missed the wood for the trees, there, Mark. blush

What I was hoping you were telling us was the Code for disabling "BAD SET" sensing. I don't see it listed in the manual. Any idea what it is? think

Joined: Jul 2010
Posts: 6
Newbie
Offline
Newbie
Joined: Jul 2010
Posts: 6
Your right Geoff i've just had a look at the latest version of the GW Manual we have got (Issue 15) and there is no mention of it.

To turn the SET TEST function off is Code 85


Mark

"Its nice to be important, but its more important to be nice"
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

That's nice information, Mark ... and I thank you for it.

OK Ed ... over to you! smile


If you don't inspect ... don't expect.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
3 members (daisizhou, David Mulvey, Huw), 4,258 guests, and 132 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5