Hi, i has problem connect MX800 to HP 2510 network switch. HP 2510 switch ports configure to 100FDx , but MX800 can show connect, some time MX800 network logo missing and need to restart MX800. hope expert here can help me this problem. thanks

Joined: Jul 2011
Posts: 16
hc1528 Offline OP
Novice
OP Offline
Novice
Joined: Jul 2011
Posts: 16
Hi, attached screen capture for MX800 interface specification. i think MX800 network working with 100mbps.

Attached Images
MX800_LAN_interface.JPG (28.29 KB, 184 downloads)
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Are you saying that it works OK some of the time - but then drops out (off the network)?

Are you sure you've got decent connections? think


If you don't inspect ... don't expect.
Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79
Hi HC,

Is the switch part of the IntelliVue Clinical Network, or part of the hospital IT network?

If it is part of the IntelliVue network, then the switch is configured wrong - it should be set to auto negotiate, or if that fails, to 10MB/s Half Duplex (definitely not 100 MB/s Full Duplex, as other components of the ICN may not support those speeds).

If your MX800 is deployed on the hospital network, you need to configure the switches as detailed in the Customer Supplied Clinical Network specifications document (requirements vary depending on the installation).

I strongly advise referring this issue to your local Philips customer service organization. A clinical network is considered part of a medical device, and any modification to such a network (which includes changing/replacing switches) must be carried out by a trained individual. A network disruption could prevent data/alarms being shown at the central station, which could lead to a delay in treatment and/or a patient incident.


Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (michal1, daisizhou), 4,915 guests, and 40 robots.
Key: Admin, Global Mod, Mod
Newest Members
Mark B, Royal, davidarhui, Czr, Jimmy Li
10,391 Registered Users
Forum Statistics
Forums26
Topics11,279
Posts74,591
Members10,391
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5