Hi Craig,
The MP50 had 2 hardware generations. Monitors with first generation main board can be upgraded to SW rev G.0.
The monitors shipping from around mid 2007 have had the second generation main board, with increased flash memory. These can be upgraded to the current SW rev, which is J.1.
The SpO2 module with Covidien Nellcor OxiMax technology is backward compatible with IntelliVue SW rev C.0.
This means, you should be able to upgrade you existing MP50s to rev G.0 (without a mainboard upgrade) and then they should be able to accept the SpO2 module with Covidien Nellcor OxiMax technology.
Please consult your Philips Field Service Engineers and they should be able to advise you on the next steps.
Chandana Samaranayake MD
Chandana Samaranayake is an employee of Philips Healthcare. All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.
As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
|