Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Oct 2013
Posts: 11
Nikoooo Offline OP
Novice
OP Offline
Novice
Joined: Oct 2013
Posts: 11
Hello all, I have a monitor Datex ohmeda S/5 light, but something he broke Noninvasive blood pressure. Pump it stopped inflating cuff. Monitor only displays "cuff loose" and does not pump the cuff. Replace hose adult with another brand new and still same problem. Does anyone know what causes this and what can I do to fix it.
Kind Regards
Nikolai
Bulgaria

Joined: Jul 2012
Posts: 6
Newbie
Offline
Newbie
Joined: Jul 2012
Posts: 6
Hello There Nikko, in many times the problem of loss pression of insufflation may be occasioned for leaks in the cuff or in the hose, if the problem persist you must check the value of pression insuflation selected in the monitor software, the last option will be check the insuflation pump within your monitor, of course you will need the service manual.

I hope can be useful my opinion.

Greetings.

Joined: Mar 2002
Posts: 188
Mentor
Offline
Mentor
Joined: Mar 2002
Posts: 188
Hi Nikoooo,
The fault is uaually the plastic chamber, with the pneumatic valves attached,inside the monitor leaking. It is formed in two halves bonded together and leaks along this join. You can sometimes find it with a little leak detector and use epoxy to seal it. If not you will need a replacement chamber.
Regards
Ed Bennett
Southwest Medical Ltd


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 15,401 guests, and 35 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,482
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5