Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#66905 25/02/14 10:51 AM
Joined: Oct 2006
Posts: 362
Likes: 37
Malcolm Offline OP
Sage
OP Offline
Sage
Joined: Oct 2006
Posts: 362
Likes: 37
Installed new battery "Physio Control 11141-00156" got half way through the PIP as per service manual, from having a full battery at the onset of the PIP it was reduced to 1/4 capacity,
only having carried out a discharge at 360J & 100J to control the discharge calibration.
Question.. anyone else has had this problem?? because this is not the first it's happened to me,maybe an inherent software problem with the PIP which over stresses the battery?
In the user manual the battery is as 12V 6.2Ah,the one supplied 12V 4.5Ah could this be some sort of an equivilent grey import,which is strange as when searching the reference number results in hundreds of apparent hits for OEM batteries

Malcolm #66911 25/02/14 10:54 PM
Joined: May 2010
Posts: 8
Newbie
Offline
Newbie
Joined: May 2010
Posts: 8
Hi Malcolm, I had this problem a couple of years ago and it turned out that I was not using the latest version of the PIP or PVP instructions. Apparently if you are using the version that asks you to do the automatic test by resetting the clock back to a date of 2004 then what happens after is it "thinks" the battery has been in use for however many years and Writes that age back to the battery which suddenly reduces the displayed battery capacity. We had several do this and one defib replaced before Medtronic forwarded us the most current version of the PIP instructions which are Titled Rev-H. In Rev -H you do NOT sent the year back in time when doing the automatic test?

Which PIP are you using?

Hope that helps

Cheers
Mike

Malcolm #66914 26/02/14 11:48 AM
Joined: Oct 2006
Posts: 362
Likes: 37
Malcolm Offline OP
Sage
OP Offline
Sage
Joined: Oct 2006
Posts: 362
Likes: 37
Mike, the next time you'r in Rome the beers are on me!!!!!
As I suspected the error was software related....
I'm using a revision from 2007, would you have the possabilty to PM me your revison H

Malcolm #66952 04/03/14 11:48 AM
Joined: Jan 2010
Posts: 73
Scholar
Offline
Scholar
Joined: Jan 2010
Posts: 73
Physio control are aware, we had this too, I always set the date to the first monday of the month for the self test test.
We also had a unit go back to physio control for the same fault, we think the date chip was faulty and would kill batteries on the first monday of the month....Please note that the newer Lifepak 12 batteries are " dated too" if you use them beyond their use by date they report as faulty.

Last edited by Mithrandir; 04/03/14 11:51 AM.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 1,843 guests, and 37 robots.
Key: Admin, Global Mod, Mod
Newest Members
james_on_ebme, Biomed Egypt, Nektarios K., Keisya Andretta, Diego Lins
10,363 Registered Users
Forum Statistics
Forums26
Topics11,253
Posts74,502
Members10,363
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5