Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 3 1 2 3
Joined: Jul 2005
Posts: 601
Philosopher
Offline
Philosopher
Joined: Jul 2005
Posts: 601
it's been a while since I last used an IDA-4 but from my recollections the IDA-4 is not accurate enough to be used for calibration purposes. I'm sure we had a "do not use for calibration" label stuck on ours. We mainly used it for long term soak tests, i.e. MS16A at 2mm/h over 16-18 hours.
I beleive that our guys follow the manufacuter's instructions for the PPM.

Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
As an aside, around 2002?, Baxter stated that the IDA was not a suitable method of accuracy determination for the flow delivery of the Colleague Volumetric Pump, and stated that its use contributed to reported inaccurate readings.

The EBME Department I was working in, were instrumental in identifying under delivery, possibly caused by shuttle mechanism drift and changing the manufacturing specification on the giving sets and raised this issue with Baxter and the MDA (as was).

But on a scheduled visit to Baxter Northampton, what did I see being used in the calibration bay by the Baxter Technicians? think

Joined: Aug 2008
Posts: 87
Likes: 6
Adept
Offline
Adept
Joined: Aug 2008
Posts: 87
Likes: 6
Using the IDA for checking purposes is perfectly acceptable I have been using my IDA 2 for many years. An experienced engineer should be able to determine from using the IDA if calibration is required, then I would be inclined to follow the recommended procedure. Going back to the point 'you have to follow the manufacturers recommendations', Nellcor 560 manual recommends the battery is changed every 2 years we all do that dont we think

Last edited by billy11; 19/03/14 9:33 AM.
Joined: Jul 2005
Posts: 601
Philosopher
Offline
Philosopher
Joined: Jul 2005
Posts: 601
Just checked the IDA manual, page 3 if you are interested. The accuracy is stated as either 1% or 2% (+/-1lsd)depending on the flowrate & test volume. It may be acceptible for calibrating volumetric pumps but isn't good enough for syringe drivers.
Burette or scales & stopwatch are best option (and cheapest!).

Joined: Sep 2001
Posts: 499
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 499
Likes: 1
Good Point BiomedBill, Alaris GH pump is a Syringe pump and Alaris GP is a Volumetric pump.....

Last edited by umish; 19/03/14 10:53 PM.

UMi-007

"WORK SMART NOT HARD !"
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
I would have thought scales are less accurate than 1% unless you have some high quality laboratory ones.
And can you read a burette to better than 1% accuracy? If you infuse a full 50ml syringe you will have to be able to read it to less than 0.5ml and be able to operate a stop watch to that accuracy as well. And remember if you use two instruments to measure anything you have to add the two inaccuracies which means you have to be able to measure each one to 0.5% to make an overall 1% apparatus.
May be the IDA is not so bad after all.
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: May 2008
Posts: 50
Scholar
Offline
Scholar
Joined: May 2008
Posts: 50
Our burettes are graduated to 0.1ml, can't speak for anyone else's of course.

In the case of the Signature and Alaris GP pumps, you calibrate according to the volume delivered rather than the time. In fact, the GP PVP doesn't actually specify a delivery time.

So far as syringe pumps go, I would have thought it better to use a linear gauge than measuring fluid delivery, since the linear guage would take out small variations in the syringes.

Joined: Sep 2001
Posts: 499
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 499
Likes: 1
The other point is you are supposed to use a calibrated giving set......


UMi-007

"WORK SMART NOT HARD !"
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Sometimes you have to take a step back and ask yourself what it is you're trying to achieve. smile

How accurate are the pumps (drivers, whatever) supposed to be? And what about repeatability of the results?

What do the specs. say? Then how accurate do they need to be (in the Real World, that is); how accurate is your test kit (or testing method), and how much time have you got? Are you just "checking", "testing" ... or "calibrating" to the n-th degree?

You also need to decide what you should be measuring (what is important); volume delivered, flow-rate, occlusion alarm pressure etc., etc. Pumps work in different ways, and some designs emphasise different aspects. And all the well-known test kit (and testing methods) have strengths and weaknesses.

Volumetric pumps and syringe pumps are doing different things. You need to understand what it is you're trying to do. Could that be the Golden Rule? think

Joined: Jul 2005
Posts: 601
Philosopher
Offline
Philosopher
Joined: Jul 2005
Posts: 601
This is why I took a step back from infusion devices. Too many conflicting views for such a simple device, added to which we have to put up with over reliance on software and never ending agruments about drug protocols and things like Guardrails etc.
A simple piece of medical equipment complicated by fear mongering (sorry, clever marketing!) and a " we've got a processor let's invent stuff for it to do whilst it is carrying out a simple task" attitude.
These were the first devices that got me into thinking that we need to temper (computer) technology by keeping an eye on the function of the device in question. Infusion devices have gone through many iterations whereby they rely more on the ability of a software programmer than the physical engineering required to make a safe, accurate & durable (emphasis on the last one) device.

Sorry to go a bit off topic, rant over.

Where did I put that burette?

Page 2 of 3 1 2 3

Moderated by  Huw 

Link Copied to Clipboard
Who's Online Now
0 members (), 12,160 guests, and 101 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,281
Posts74,584
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5