If I remember rightly, HEI-95 (and DB9801 Sup 1) went for 1 Amp. Yes; other historic documents are available (and more recent "guidance" has been silent about specific test current values)! And as Malcolm has indicated, various well-known bits of test kit (having been introduced at various
epochs in the history of medical electrical safety) have tended to "do their own thing" (or perhaps simply followed what was
de rigueur at the time).
Myself, I have usually thought that 10 Amps for five seconds was a good (and, dare I say, Real World) test current. Whilst I can see the merit in frying a few strands with 25 Amps, I can't really support that idea whilst also preaching that 10 Amp fuses be fitted to the ubiquitous IEC mains cables that are also rated at 10 Amps!
But rather than lay it all out again ... anyone who's still interested in this stuff could do worse than re-visit some of the many earlier threads that have touched on this issue;
this one, for example.

See also John's
article.
Meanwhile, what does 62353 actually say about the PE test current?