We recently had our intensive care ward refurbished, on completion I noticed the wall mains sockets(housed in a metal case) on some bedsides have now an equipotential stud.....any idea why they are there!!!!!!!!!!!!!!!!


Failure to Prepare is to Prepare for Failure
#6812 12/07/04 12:28 PM
Joined: Jun 2003
Posts: 338
Master
Offline
Master
Joined: Jun 2003
Posts: 338
It is so you can plug your Electrostatic wristband in and work on the equipment at the patients bedside. - NHS Efficiency measures. LOL laugh


It is better to be reactive than radioactive...
#6813 12/07/04 1:16 PM
A
Anonymous
Unregistered
Anonymous
Unregistered
A
So that you can connect a secondary earth lead from a medical device to the earth socket.

See IEE Wiring Regs BS 7671 Guidance Note 7 Special Locations, there is a section for Medical Locations. The installation should have also been of an isolated type (IT System) with appropriate monitoring.

See http://www.bender-uk.com/www/bendergmbh/bendergmbh.asp for isolated supply equipment/installations.

Bender UK Ltd
Low Mill Business Park
Ulverston
Cumbria
LA12 9EE

tel 01229 480 123
fax 01229 480 345

Regards,

Mark

#6814 12/07/04 3:13 PM
Joined: May 2003
Posts: 27
Sav Offline OP
Dreamer
OP Offline
Dreamer
Joined: May 2003
Posts: 27
Thanks Mark.................but why would you want to connect a secondary earth lead from the device to Earth........
Would it also be there to reduce interference????


Failure to Prepare is to Prepare for Failure
#6815 13/07/04 2:36 PM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71
Sav,

I my memory serves me well, the “equipotential terminal” was a feature of DIN (German) standards relating to “rooms intended for clinical use”. They were useful in, for instance, ECG recording areas (stress testing set-ups would be an example) by providing a “central earthing point” for all significant equipment in the room.

Connections to equipment would be made using temporary earthing cables, with clips or press-stud ends. Electrically, then all equipment casings would be bonded together at the equipotential terminal (using daisy-chain, or star, or whatever gave best results). Technically, this point would be different from protective earth (although you would expect the two to be at the same potential – ie, “earth”). Using this set up could sometimes eliminate electrical interference, although I expect the original idea was to do with electrical (patient) safety as we now know it. Remember that back in the “good old days”, electromedical equipment in some parts of the world sometimes worked better with an earthing wire running to a clip on the plumbing!

Note that a similar “stud” is found on the back panel of quite a few bits of electromedical kit intended for the IEC or “international” markets.

I am surprised that equipotential terminals have been installed in a new refurbishment (maybe there’s a new requirement that I don’t know about – quite likely). I expect the clue is in the fact that you mention that the outlets have metal cases. If I were you I would simply regard it as a “belt and braces” feature that may be handy when carrying out electrical safety tests of the outlets concerned and of any equipment nearby. Bottom line is, no need to worry.


If you don't inspect ... don't expect.
#6816 14/07/04 2:16 PM
Joined: May 2003
Posts: 27
Sav Offline OP
Dreamer
OP Offline
Dreamer
Joined: May 2003
Posts: 27
Thanks Geoff Hannis.......much appreciated.
Cheers


Failure to Prepare is to Prepare for Failure
#6817 16/07/04 9:30 AM
Joined: May 2003
Posts: 75
Adept
Offline
Adept
Joined: May 2003
Posts: 75
my understanding is these studs are being installed to reduce the chance of micro-shocks which can be caused by slight difference in earth potentials.when catheters ect puncture the skin the resistance to shock is greatly reduced.I think the limit is 20 millivolt between equipment,if this limit is higher it needs to be connected to the equipotential point via a seperate lead.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 1,989 guests, and 18 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,250
Posts74,485
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5