I can well imagine that the Alternative Method is not easily understood (and I wonder how instructors on electrical safety courses get on in that regard).
In fact some may even go as far as to suggest that it's a bit of a "fudge"!
Although I can see the attraction to using the Alternative Method (in short:- "quick and easy"), I would argue myself that it doesn't actually measure leakage currents at all, but rather
estimates (and usually an
over-estimation at that) what they might be!
Biomeds of the type who prefer things like EST to be Step 1), Step 2)
etc. (that is,
simplified) may not be too happy about having to choose between the three approaches ("Methods") that 62353 supports - and my guess is that the (more traditional) Direct Method is the one most techs will be following (as it is easily understood, and "follows on" from what folk have been used to - in other words, 60601).
Lastly, it is my understanding (?) that in cases where the Alternative Method indicates currents around the edges of acceptable limits (or above them), then 62353 suggests using another Method (
eg, Direct) to confirm them!!
Maybe we need a Poll (and we haven't seen one in a while) to find out what the various Departments
are doing when it comes to 62353.
BTW: have the Standards Committee(s) made up their mind yet about what the Protective Earth Test Current should be?