Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Apr 2002
Posts: 135
PaulKWJ Offline OP
Expert
OP Offline
Expert
Joined: Apr 2002
Posts: 135
Hi All

We have a couple of Sensormedics 3100B and we have a reoccurring problem of the pressure line, all the way up to the pressure transducer, getting contaminated with fluids.

Does anyone else suffer with this kind of issue with the 3100B? and did you ever resolve it?

Many thanks in advance.


New beginning. New location. New adventure. Old grumpy me!
Joined: Mar 2013
Posts: 22
Dreamer
Offline
Dreamer
Joined: Mar 2013
Posts: 22
If the machine is set up properly this is not an issue, the pressure line has a constant purge flow to stop this. This flow is checked on the service and should equate to a pressure of 140cmH2O when checked with the airway pressure line clamped. This is adjusted with PR6.

This should avoid the issue you are having. The only other way this happens is if the circuit is set up incorrectly and the drain line is connected to the airway pressure port.

Joined: Apr 2002
Posts: 135
PaulKWJ Offline OP
Expert
OP Offline
Expert
Joined: Apr 2002
Posts: 135
Hi Darkmarker

Thank you for your reply.
Yes, the manometer purge line pressure does get checked every 6 months and tweeked if required.
The clinical staff are trained to set the circuit up and complete the Pre-Use checks. They suggest the machine is just faulty every time this event occurs.


New beginning. New location. New adventure. Old grumpy me!

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 1,363 guests, and 260 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,280
Posts74,583
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5