|
|
|
|
Joined: Dec 2016
Posts: 13
Novice
|
OP
Novice
Joined: Dec 2016
Posts: 13 |
Dear All,
A Braun perfusor fm (MFC) syringe pump, a fault is displayed as a SYRINGE ALARM after a few seconds that was switched ON. Anyone can help me on this issue please?
Best Regards, Techice.
|
|
|
|
|
Joined: Jul 2002
Posts: 2,020
Hero
|
Hero
Joined: Jul 2002
Posts: 2,020 |
Frank has the service manual, that might help you. Robert
My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
|
|
|
|
|
Joined: Dec 2016
Posts: 13
Novice
|
OP
Novice
Joined: Dec 2016
Posts: 13 |
Hi ROJo, Thanks for your reply.Alreday have the service manual. Have this manual,Please see the below link: http://www.frankshospitalworkshop.com/eq...vice_manual.pdfBut regards this error SYRINGE ALARM,I cant find any explanation. Best Regards, Techice
|
|
|
|
|
Joined: Feb 2004
Posts: 14,802 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,802 Likes: 72 |
It gets an honourable mention in passing on page 2-11.  To be honest, compared with some (many) "modern" manuals I have seen, that one appears quite comprehensive.
|
|
|
|
|
Joined: Dec 2016
Posts: 13
Novice
|
OP
Novice
Joined: Dec 2016
Posts: 13 |
Hi Lee,
Thanks for your reply,
There is no information on page 12,(A fault is displayed as a SYRINGE ALARM)I would be much appreciated if someone tell me how a drive unit(Syringe Plunger)alignment can be done.
I am 100% sure that the fault comes from the drive unit(Syringe Plunger).
Techice
|
|
|
|
|
Joined: Dec 2016
Posts: 13
Novice
|
OP
Novice
Joined: Dec 2016
Posts: 13 |
Hi Geoff,
Thanks for your reply.
Sorry,but I read all the pages on service manual and user manual but no information found regards SYRINGE ALARM error.
I am 100% sure that the problem comes from the drive unit(Syringe Plunger)do you know if there is any alignments what can I do?
Techice
|
|
|
|
|
Joined: Sep 2006
Posts: 745
Philosopher
|
Philosopher
Joined: Sep 2006
Posts: 745 |
Hello Techice
Page 12 says under Operating alarms
"Syringe alarm Fault on syringe replacement? Manipulation of syringe holder during infusion?"
So presumably you need to be looking at what ever systems monitor that a syringe is in place and possibly the sensor for detecting the size of the syringe.
Lee
Don't forget "we've never had it so good".
|
|
|
|
|
Joined: Dec 2016
Posts: 13
Novice
|
OP
Novice
Joined: Dec 2016
Posts: 13 |
Hello Lee,
Thanks for your reply.
Yes, I see this notice as well. But not the syringe holder for sure. I replaced with another drive unit(Syringe plunger)and the syringe pump works well.
In fact I perform a syringe holder(Syringe clamp)calibration to make sure that works good.
I think something in the drive unit(Syringe plunger)like a calibration or alignment.
Best Regards, Techice
|
|
|
|
|
Joined: Sep 2006
Posts: 745
Philosopher
|
Philosopher
Joined: Sep 2006
Posts: 745 |
Hello Techie
On the Perfusor Space version there is a sensor in the drive head that detects contact with the syringe plunger and the position of the claws is also monitored.
If the FM is the same; then your fault could have been either of those.
Lee
Don't forget "we've never had it so good".
|
|
|
|
1 members (daisizhou),
3,150
guests, and
35
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,254
Posts74,506
Members10,364
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|
|
|