Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 3 1 2 3
Joined: Feb 2007
Posts: 303
Likes: 17
MikeX Offline OP
Master
OP Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
Currently BS 7671 specifies the quantity of equipotential connection points required for medical locations. There are proposals to change this in the next version to allow the designer, in conjunction with the medical staff, to decide on the quantity and even if to supply them at all in some areas.

Are the connection points actually used to connect medical equipment, via the equipotential bonding connector sometimes found on medical equipment, to the supplied connection point of the electrical installation?

Do you use equipotential connection leads with medical equipment
single choice
Votes accepted starting: 30/08/17 7:56 AM
Joined: Feb 2007
Posts: 303
Likes: 17
MikeX Offline OP
Master
OP Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
If someone votes they use them occasionally could they post why they have used them and for what type of equipment.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

"They" (the cardiographer) thought that using* them might reduce the "interference" on the electrocardiogram. It didn't.

But what did improve the trace was my explanation that, nine times out of ten, such problems were (are) due to "poor" contact at the patient electrodes.

Bottom line? IMHO Equipotential Points are (yet another) Red Herring (at best); and waste of time and money (at worst). frown

As an aside, I have also seen them put to good use in providing a de-facto "earth" for equipment such as CTG, ECG, EEG and similar where, for whatever reason (the "wrong" mains cable and [or] plug being the usual one), no integral protective conductor was in place. But, I hasten to add, such instances have been "outside of the UK"!

* Stringing a green/yellow coloured cable between the stud on the ECG Recorder to the stud on the wall.


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Although nothing to do with providing them within a medical location (aka room) - they* often come in handy when trying to find an "earth" point for carrying out electrical safety tests. Allegedly. whistle

* The Equipotential Stud on the back of some machines.


If you don't inspect ... don't expect.
Joined: Feb 2017
Posts: 24
Likes: 1
Dreamer
Offline
Dreamer
Joined: Feb 2017
Posts: 24
Likes: 1
Geoff totally agree with you

Joined: Oct 2006
Posts: 364
Likes: 38
Sage
Offline
Sage
Joined: Oct 2006
Posts: 364
Likes: 38
Also very handy in Italy for removing the ever present static charge.

Joined: Feb 2007
Posts: 303
Likes: 17
MikeX Offline OP
Master
OP Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
Originally Posted by Malcolm
Also very handy in Italy for removing the ever present static charge.
On what equipment?

Class I medical electrical equipment should have a functioning earth connection anyway. Class II devices are rarely fitted with a connection point. Patient beds and other such equipment will have anti-static wheels (or other similar measures), as they are intended to be moved.

As most static is generated by people with incorrect (not anti-static) clothing, floors that are not anti-static or the humidity is too low. I have seen it in many rooms where the cleaner uses the wrong polish and creates a high resistance barrier leading to build up of static. This build up can play havoc with ECG traces never mind static shocks. No amount of equipotential bonding connections will help here!

I should also point out that these connections they were never intended to reduce static!

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Malcolm, get rid of that Bticino stuff ... and get some proper plugs and sockets in there (eg, BS-1363). smile

Meanwhile, here are a couple of nice (but oldish) .pdf's that may help in understanding the issues involved:-

1) electrical.theiet.org/wiring-matters/38/medical-locations.cfm?type=pdf
2) electrical.theiet.org/wiring-regulations/updates/gn7-2015-errata.cfm?type=pdf

To "cut to the chase", Equipotential Bonding is really about earthing the water taps, gas pipes and so forth! whistle

In other words, it is really to do with the building, rather than the portable medical equipment per se.


If you don't inspect ... don't expect.
Joined: Sep 2017
Posts: 111
Likes: 3
Savant
Offline
Savant
Joined: Sep 2017
Posts: 111
Likes: 3
Good evening. If it helps, supplementary bonding conductors are meant to act as a back up to the medical device's earth conductor in the mains flex. It serves no other purpose. The connection points you refer to I guess are the supplementary bonding connectors referred to in 9.12 in GN7? The wall connectors should not be used to measure the earth resistance in a medical location. The reference point to be used is the Equipotential Bonding Busbar in the medical location. However, the SBC can at least be used for quick checks.


Regards

Ian

Last edited by Huw; 04/09/17 8:06 PM. Reason: Removed link.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Welcome to the forum, Ian. smile

This brochure makes interesting reading.

Note the stated requirements of (the present edition of) BS-7671.

My own "doubts" about the value of Supplementary Equipotential Bonding plugs and sockets in medical locations is that they are just something else for medical staff to worry about (misunderstand, whatever). And they have, by the way, been about in some hospitals for many years ... so I can attest to the "confusion" clause!

I also believe there may also sometimes be a level of confusion regarding the usefulness (or otherwise) of SEB (SBC, whatever) in medical rooms, and the related Equipotential "studs" often found on the back of portable and (or) mobile medical equipment.

This last category (the studs) may have some value in certain situations (cath lab, major surgery etc.) where "daisy chaining" together the studs of many items of equipment to ensure that all equipment chassis sit at the same zero volts may be beneficial for various reasons (the main one being, of course, the electrical safety of the patient).

Meanwhile, I continue to uphold the view that the primary "safety feature" of Class I medical equipment is the protective ("earth") conductor in its mains power cable - and that safety (and a whole load of other Good Things) is ensured by a continuous programme of properly conducted preventive maintenance - by the biomeds. In fact it has always be thus! That's where the money needs to be spent; not on ever increasing layers of "smart" technology. smile


If you don't inspect ... don't expect.
Page 1 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 8,495 guests, and 292 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,285
Posts74,589
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5