Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 3 of 4 1 2 3 4
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Originally Posted by Geoff Hannis

In other words, how accurate are these predictions?


See here. smile


If you don't inspect ... don't expect.
Joined: Sep 2010
Posts: 412
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 412
Likes: 2
..what the use of test instrument by the way?


--- PATIENT SAFETY FIRST ---
Joined: Feb 2009
Posts: 1,908
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,908
Likes: 18
Healthcare institutes must be single phase, dual phase was, and still is the lazy way to upgrade from 110V to 220V


I am not Flippant, I am Smart
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Eh? What about 3-phase (for x-ray systems, etc.)? think

I think what you meant to say there, Neil, was that (in some locations) both 110 and 220 VAC may be taken from a supply transformer in star configuration!

But (and perhaps more importantly in such cases) where is the earth connected?

I have always thought that "two-phase power supplies" was an "American idea". I have only ever encountered them in buildings built to "domestic standards" in places like Saudi Arabia (village clinics, and the like). Only once did I come across a piece of "medical equipment" that actually required a 220 VAC two-phase supply in a proper hospital, and that was an ancient baby milk-bottles sterilizer (I think it was) made in the USA.


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Originally Posted by ariel dulay

..what the use of test instrument by the way?


Do you mean the 62353 Alternative Method?

For more, see this earlier thread.

You will need an Safety Tester that can accommodate this (has this "feature").


If you don't inspect ... don't expect.
Joined: Nov 2007
Posts: 117
Savant
Offline
Savant
Joined: Nov 2007
Posts: 117
Going back with the initial queries. In our policy in Malaysia, when we performing PPM inclusive with EST for any medical equipment regardless class 1 or class 2, it shall be isolated from patient. This shall minimize any potential risk exposed to the patient which will lead into severe damage.

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

OK, that's the policy. But how do you implement the policy in cases where beds are hardly ever (if at all) vacated?

Are you following 62353 ... and especially, the Alternative Method? think


If you don't inspect ... don't expect.
Joined: Nov 2007
Posts: 117
Savant
Offline
Savant
Joined: Nov 2007
Posts: 117
We never experienced that.

Normally bed, the frequency for the PPM is once a year and bed hardly vacated is normally in ICU.

We still emphasize in 60601 and not yet 62353.

Joined: Feb 2009
Posts: 1,908
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,908
Likes: 18
For ICU, ask the head nurse to inform BME when the room becomes vacant, death, theatre procedure or any other service outside of the ICU and PPM all the equipment for that room.


I am not Flippant, I am Smart
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

1) Patient goes to Theatre
2) Nurse informs BME
3) BME rushes to ICU
4) Bed fails EST (PPM)
5) Bed has a fault
6) Bed needs extra work
7) Bed needs parts
8) Patient returns
9) What happens next?


If you don't inspect ... don't expect.
Page 3 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 5,424 guests, and 18 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,358
Most Online53,260
Apr 28th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5