|
|
Joined: Feb 2009
Posts: 1,909 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,909 Likes: 18 |
Geoff, all depends on the quality of the incoming water, un-used water from the loop is returned to the system via the hot water tank. Hopefully in Arials case I hope that they are not using UV as a form of sterilization as this is a major source of endotoxins.
I am not Flippant, I am Smart
|
|
|
|
|
Joined: Feb 2004
Posts: 14,814 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,814 Likes: 72 |
As an aside, I have (had) always found the water treatment side of dialysis very interesting; although my own involvement (a very long time ago now, Neil - and even before we met) was in support of a number of small AKU's in desert locations. I recall lots of pump failures and so forth (simple stuff, but inexcusable really). As you must remember, the incoming water supply from the baladiya was often of "poor" quality and sometimes intermittent; I seem to remember on occasion having to bring in tankers (trucks) full of so-called "sweet water". All a bit ad-hoc perhaps, but it was very much "needs must" at the time. I don't remember any UV sterilization in those AKU's, though. If I remember rightly (and I may still have my old notebook - the pen and paper type, that is) it was all done via Reverse Osmosis (Permeator?); but I seem to recall "salts" (softener?) being involved as well. Also a storage tank, of course - and a 5 micron (I think it was) filter before each of the dialysis units. All a bit crude back then (and there), no doubt. 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Feb 2009
Posts: 1,909 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,909 Likes: 18 |
Geoff, after all the softeners, Filters and the RO some people thought it would be wise to incorporate a UV filtration system to the output of the RO, thus killing any residual bacteria but creating endotoxins. You can still see diagrams on google with this type of filtration.
I am not Flippant, I am Smart
|
|
|
|
|
Joined: Sep 2010
Posts: 413 Likes: 2
Sage
|
OP
Sage
Joined: Sep 2010
Posts: 413 Likes: 2 |
Geoff, after all the softeners, Filters and the RO some people thought it would be wise to incorporate a UV filtration system to the output of the RO, thus killing any residual bacteria but creating endotoxins. You can still see diagrams on google with this type of filtration. Boss Neil kindly explain further why you mention that UV filtration can cause endotoxins. For curiuosity, are UV light that installed at the end of the R.O. system and UV filtration that you mention are the same?
--- PATIENT SAFETY FIRST ---
|
|
|
|
|
Joined: Sep 2010
Posts: 413 Likes: 2
Sage
|
OP
Sage
Joined: Sep 2010
Posts: 413 Likes: 2 |
HEMODIALYSIS (HD) VS HEMODIAFILTRATION(HDF)
- any idea? - what are the dis advantage and advantage?
--- PATIENT SAFETY FIRST ---
|
|
|
|
|
Joined: Feb 2009
Posts: 1,909 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,909 Likes: 18 |
Clinical question not engineering
I am not Flippant, I am Smart
|
|
|
|
|
Joined: Feb 2009
Posts: 1,909 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,909 Likes: 18 |
"Boss Neil kindly explain further why you mention that UV filtration can cause endotoxins.
For curiosity, are UV light that installed at the end of the R.O. system and UV filtration that you mention are the same?"
They are the same, and we don't install them. Google endotoxins as there is a large amount of information. If your RO plant has a UV filter at the output then you should upgrade.
I am not Flippant, I am Smart
|
|
|
|
|
Joined: Aug 2005
Posts: 14
Novice
|
Novice
Joined: Aug 2005
Posts: 14 |
Hi Ariel A very good website for discussion forum and all things renal is: https://www.renaltech.netBest wishes Roger
|
|
|
|
|
Joined: Sep 2010
Posts: 413 Likes: 2
Sage
|
OP
Sage
Joined: Sep 2010
Posts: 413 Likes: 2 |
Thanks all of you Folks your input is highly Appreciated..
--- PATIENT SAFETY FIRST ---
|
|
|
|
1 members (daisizhou),
1,823
guests, and
16
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,278
Posts74,590
Members10,391
| |
Most Online69,877 Jul 17th, 2026
|
|
|
|