Hi Guys
Has anyone come across an alternate tester for testing patient scales.
My hospital has 50 odd scales and I have lost the strength to be able to wheel around 250 kgs for my annual testing . Something like a calibrated strain gauge or ...
Cheers
|
|
|
|
Joined: Feb 2004
Posts: 14,664 Likes: 62
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,664 Likes: 62 |
Use plastic containers and water ... using the well known formula that relates volume of water to weight (mass).
If you don't inspect ... don't expect.
|
|
|
|
Joined: Oct 2006
Posts: 325 Likes: 28
Master
|
Master
Joined: Oct 2006
Posts: 325 Likes: 28 |
Andy, Sounds like you need an apprentice !!
|
|
|
|
Joined: Sep 2005
Posts: 67 Likes: 6
Scholar
|
Scholar
Joined: Sep 2005
Posts: 67 Likes: 6 |
We don't check the calibration of scales, however, what would be wrong with just weighing yourself on a calibrated scale then quickly going round and weighing yourself on the 50 odd scales you have and if the measurement ties up(within a few percent) it passes.
|
|
|
|
Joined: Feb 2004
Posts: 14,664 Likes: 62
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,664 Likes: 62 |
See this earlier thread ... and this one.
If you don't inspect ... don't expect.
|
|
|
|
Joined: Aug 2008
Posts: 12
Novice
|
Novice
Joined: Aug 2008
Posts: 12 |
I can see the auditors having a field day with all these suggestions, the only correct way is to have 3 calibrated reference weights and use them. You may break your back but you are covering your backside.
|
|
|
|
Joined: Feb 2004
Posts: 14,664 Likes: 62
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,664 Likes: 62 |
Which auditors would that be? And how do you calibrate weights? Certify, may be. And what tolerance is allowed?
If you don't inspect ... don't expect.
|
|
|
|
Joined: Feb 2009
Posts: 31 Likes: 4
Visionary
|
Visionary
Joined: Feb 2009
Posts: 31 Likes: 4 |
I have a compression jig that can measure up to 500KG - use a car jack to make the compression.
The accuracy can easily be established by using known controlled weights and measuring main scale with these - I currently have accuracies of 1g in 100Kg - not high level, like lab level scales, but OK for human scales.
The whole thing weighs about 40 KG, and when its used correctly - you can measure any digital scales that lock out after a period of time, ie I regularly test Seca 813's to their max weight of 200Kg.
I believe all scales ought to be weight check across 100% of their weight range to ensure there is no mechanical sticking or binding at some point.
Phil
|
|
|
|
Joined: Feb 2009
Posts: 1,893 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,893 Likes: 18 |
Phil, do you have a calibration certificate for your 'Jig' if not then all data is invalid.
JCI, CBAHI, ISO etc. require that all test equipment must be calibrated and certified annually.
If you think hiring professionals is expensive, try hiring amateurs!
|
1 member likes this:
iozzie |
|
|
|
Joined: Feb 2009
Posts: 1,893 Likes: 18
Hero
|
Hero
Joined: Feb 2009
Posts: 1,893 Likes: 18 |
Weights need a calibration certificate on an annual basis the same as test equipment
If you think hiring professionals is expensive, try hiring amateurs!
|
1 member likes this:
iozzie |
|
|