Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Oct 2011
Posts: 279
Likes: 22
Master
OP Offline
Master
Joined: Oct 2011
Posts: 279
Likes: 22
Conmed Sabre 2400 high frequency electrosurgical knife HLP-5.

!!!!!!!This is not caused by an error!!!!!!!!!!!!!!!!!!!!!

Because the electric knife has been used for a long time.I used an electrosurgical analyzer for energy analysis and found the energy to be inaccurate.So I did an energy calibration according to the service manual.
The calibration of pure cut, mixed cut, coagulation and bipolar mode was carried out separately
When I did all the calibration steps, I got the HLP-5 error after POST.
So I restored the program of the U4 eeprom on the cpu circuit board before the calibration, the HLP-5 error did not show up.Can enter the program normally and complete the self-test.

I don't know why, if you have the same situation as me please tell me the reason, thanks for your help!


[Linked Image]

Last edited by daisizhou; 27/01/22 2:22 PM.

daisizhou#sina.com (#=@)
Joined: Sep 2003
Posts: 4
Newbie
Offline
Newbie
Joined: Sep 2003
Posts: 4
Hi

The HLP-5 error is indicating a fault within the RF amplifier. Check all the base and collector fuses for failures as a starting point. If a fuse is open it usually means the corresponding output transistor has failed.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 1,398 guests, and 43 robots.
Key: Admin, Global Mod, Mod
Newest Members
Abdulaziz abed, Fernando Henz, Nick.A, Tdayve, AdrianT82
10,373 Registered Users
Forum Statistics
Forums26
Topics11,262
Posts74,527
Members10,373
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5