Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
Dear Gentelman / Madam,
I have Philips Avalon FM20 device to repair. It is resseting continuously just after turnig on.
I checked power supply, interface board, cabling. Seems something with main PCB.
Does someone have realted experience to share?

Joined: Nov 2012
Posts: 9
Newbie
Offline
Newbie
Joined: Nov 2012
Posts: 9
Hello,

Yes, you are correct the main cpu needs to be replaced and reconfigured via the intellivue support tools

Thanks

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72
Does this one have the battery option? ... if so, I would take a look around there.

Otherwise, this sounds like a problem that often afflicts laptops. But you're saying that the power supply is OK. Are you sure that the voltage doesn't drop down (or out) when starting up?

Myself, I doubt that major components are at fault ... more likely to be loss of battery-backed start-up parameters (NVRAM, or whatever) - or even dust or (and) fluff!


If you don't inspect ... don't expect.
Joined: Oct 2011
Posts: 284
Likes: 24
Master
Offline
Master
Joined: Oct 2011
Posts: 284
Likes: 24
I have encountered the same failure, you need professional tools.
If you need more detailed information, please contact me by email


daisizhou#sina.com (#=@)
Joined: Nov 2010
Posts: 60
Scholar
Offline
Scholar
Joined: Nov 2010
Posts: 60
Hi
First option;

Perform a cold restart in service mode.
In menu settings - Main Setup-Hardware-Coldstart.

Second option;

Check very carefully the motherboard with a magnifying glass.
Some monitors have this problem because
The hot glue that the electrolytic capacitor is glued has dissolved the conductors and contacts on the motherboard.
After the restoration of contacts, the performance is restored smile

Joined: Oct 2011
Posts: 284
Likes: 24
Master
Offline
Master
Joined: Oct 2011
Posts: 284
Likes: 24
No, the usual repair methods do not apply to FM20 repairs.
You can check the manufacturing process of the PCB circuit board, it is almost perfect, there are no manufacturing defects or similar problems


daisizhou#sina.com (#=@)
Joined: Nov 2010
Posts: 60
Scholar
Offline
Scholar
Joined: Nov 2010
Posts: 60
Hi,
Corrosion of the main board processor as a result of the use of hot (gun) glue.
I'm sorry didn't take pictures now I think I should have smile
1.Avalon FM20,s/n DE53023529
In 2019 did the repairs myself.
2.Avalon FM50 s/n DE81106025
In 2020 the board from this monitor was sent to the manufacturer.

Joined: Oct 2011
Posts: 284
Likes: 24
Master
Offline
Master
Joined: Oct 2011
Posts: 284
Likes: 24
You mean that yellow hot melt glue?
The phenomenon of the author on the first floor, I think it is more like repeated restarts caused by the loss of configuration information


daisizhou#sina.com (#=@)
Joined: Nov 2010
Posts: 60
Scholar
Offline
Scholar
Joined: Nov 2010
Posts: 60
I do not remember the color of the compound with which the capacitor is fixed smile
Сorrosion began just under it, as a result, electronic tracks were dissolved, but it had a much stronger effect on the via metalized holes.
The second (FM50 )monitor was released in 2012, in 2020 the board was destroyed so much that it was not possible to restore it.
So we sent it back to the manufacturer.
P.S.
By the way
Very little information from the topic starter's
But can step by step, if they are present, disable SPO2,NIBP,Lan, interfaces and recorder.
Pay attention to the LED power-on indicator during the loading phase.

Last edited by SeniorJunior; 22/07/22 2:55 PM.
Joined: Oct 2011
Posts: 284
Likes: 24
Master
Offline
Master
Joined: Oct 2011
Posts: 284
Likes: 24
Is it this board?

Attached Images
1853730833.jpg (97.15 KB, 26 downloads)
232322187.jpg (92.13 KB, 26 downloads)

daisizhou#sina.com (#=@)

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (fanie), 4,901 guests, and 44 robots.
Key: Admin, Global Mod, Mod
Newest Members
Czr, Jimmy Li, AndrewS1, Lam, Danny1621
10,388 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,579
Members10,388
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5