Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#76647 30/01/22 10:27 PM
Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
Hi,
Does someone know, where I can buy calibration kit for Carefusion, Fresenius, Braun pumps in EU/UK?

Joined: Oct 2006
Posts: 362
Likes: 37
Sage
Offline
Sage
Joined: Oct 2006
Posts: 362
Likes: 37
Good morning Michal, I think the biggest problem would be sourcing the propriety software needed to reset the

service interval timer after completing the calibration, in the case of BBraun (and I assume others) you need to be

an authorised agent.

What calibrations are you needing to carry out ??

Malcolm.

Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
I have alaris pump that needs syringe size calibration, and several Fresenius Agilia pumps that require at least calibration check due to service manual. You can reset service interval date is this pumps. There are alaris calibration tools available in US, put I not sure if they are going through customs (CE, declaration of confirmity)

Joined: Oct 2006
Posts: 362
Likes: 37
Sage
Offline
Sage
Joined: Oct 2006
Posts: 362
Likes: 37
Alaris

Ouch !!!!

Joined: Aug 2011
Posts: 5
Likes: 5
DTP Offline
Newbie
Offline
Newbie
Joined: Aug 2011
Posts: 5
Likes: 5
The Carefusion devices, the calibration was done through the pump. Have you tried contacting Carefusion sales to see if they will sell to you. At a stretch; if you just need syringe size calibration you could try using the relevant size BD syringe? Verify diameter with calipers.

B.Braun, you do need the correct version of HiBaSeD to go with your device firmware version to complete calibration. Verification is completed on the pump but you'll need their physical service key to fully complete 'Technical Safety Check' (TSC). They may sell you the hardware but as Malcolm has said, you need to be registered with them to be able to access their software downloads.

I have no maintenance experience of the Fresenius other than Pilote pumps - sorry.


A man may fight for many things. His country, his friends, his principles, the glistening tear on the cheek of a golden child. But personally, I'd mud-wrestle my own mother for a ton of cash, an amusing clock and a sack of French porn.
1 member likes this: Huw
Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
Yes, I tried calibrating on syringes with similar size, but it is not as precise as calibration tool - alaris.
I saw offer on partsource and despite of price level, I not sure if it can pass EU customs. Alaris requires calibration on 10,15,20,25,30mm due to service manual. I will try to manufacture in on my own.
Alaris representative in here is not going to sell service tools, even if price will be much higher than on partsource.


Fresenius Agilia syringe pumps also requires service tool - force tool, displacement, syringe size.

Joined: Sep 2009
Posts: 23
Dreamer
Offline
Dreamer
Joined: Sep 2009
Posts: 23
Hi, I can help you if you want. I am looking for braun software, for the newer version.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 24,626 guests, and 31 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,482
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5