I tried to make a self print on a sony up-21MD printer, by pressing left and right arrow while powering up, but I got this message > DESIGNER MODE>, I restart the printer but same message, how come back to the main screen?
Geaoff thank for the service manual, but noting also about DESIGNER MODE, I tried the reset with Stop, Capture and print buttons, but nothing? Fater power off and on, same message. The printer was working properly.
Aha ... so <-P<- becomes VPR. Well, VPR is the name of the top board. The other display is obviously a corruption (could just be a bad connection somewhere - how about at the display?).
As I've run out of ideas, here's a final shot:- somehow (perhaps by pressing some sort of random key-press sequence) you have stumbled across an undocumented mode (that is, "Designer Mode").
What is the condition of the unit (inside)? Clean, or not so clean?
I would try cleaning where needed, and re-seating all boards and connectors.
I don't suppose you have access to another of these printers?
Seems there is a problem with the LCD MOUNTED CIRCUIT BOARD, KEY-49, there was not the MENUE Button, we replaced it with another button but not the same, Think the comes to this board, now the lanagage is it Italian or portuguese don't know.
We will try to find this circuit board and replace it.