Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
I tried to make a self print on a sony up-21MD printer, by pressing left and right arrow while powering up, but I got this message > DESIGNER MODE>, I restart the printer but same message, how come back to the main screen?

Last edited by Aisseke; 11/09/23 5:52 PM.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
I don't recall ever seeing <DESiGNER MODE>, but Frank has the user manual.

You could try:-

1) Switch printer off.
2) Switch printer on.
3) Press and hold left and right arrow keys and MENU button simultaneously.


If you don't inspect ... don't expect.
Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
Geoff I tried this action, but I still got the same message and under it VPR. Don't reallly know the meaning

Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
Deoff, there is nothing about this message in the user manual

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
You may need the Service manual.


If you don't inspect ... don't expect.
Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
Geaoff thank for the service manual, but noting also about DESIGNER MODE, I tried the reset with Stop, Capture and print buttons, but nothing? Fater power off and on, same message. The printer was working properly.

See image

Attached Images
20230912_113543.jpg (55.11 KB, 44 downloads)
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
As mentioned before, I don't recall ever seeing DESiGNER MODE ... and don't see it in the manual either.

But ... do you read that message as DESiGNER MODE, then?

Plus the <-P<- indicator (which may be a clue); could this mean "Press the Memory Page button", I wonder (just a guess)?

I'm also wondering if the contents of EEPROM (U102?) have become scrambled.

Am I right in thinking you cannot attain Service Mode at all?


If you don't inspect ... don't expect.
Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
Sometimes it's written DESIGNER MODE (see picture sometimes) DESIOOER OODE, See attached when I wan to perform a mechanical adjustement is weird.

When I pressed on Memory page button nothing happened.

You right I cannot attend service mode at all

Attached Images
20230912_134708.jpg (67 KB, 24 downloads)
20230912_140555.jpg (57.7 KB, 23 downloads)
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Aha ... so <-P<- becomes VPR. Well, VPR is the name of the top board. The other display is obviously a corruption (could just be a bad connection somewhere - how about at the display?).

As I've run out of ideas, here's a final shot:- somehow (perhaps by pressing some sort of random key-press sequence) you have stumbled across an undocumented mode (that is, "Designer Mode").

What is the condition of the unit (inside)? Clean, or not so clean?

I would try cleaning where needed, and re-seating all boards and connectors.

I don't suppose you have access to another of these printers?

Good Luck (and fingers crossed).


If you don't inspect ... don't expect.
Joined: Sep 2023
Posts: 12
Aisseke Offline OP
Novice
OP Offline
Novice
Joined: Sep 2023
Posts: 12
Seems there is a problem with the LCD MOUNTED CIRCUIT BOARD, KEY-49, there was not the MENUE Button, we replaced it with another button but not the same, Think the comes to this board, now the lanagage is it Italian or portuguese don't know.

We will try to find this circuit board and replace it.

Thanks for your support Geoff

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 10,403 guests, and 52 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5