Joined: May 2011
Posts: 27
Dreamer
|
OP
Dreamer
Joined: May 2011
Posts: 27 |
Hi All
Just about to start an evaluation of Thermometers...... I have only ever had the privilege of working with one type of thermometer so would really appreciate your input. Just wondering from a service perspective what issues have you encountered and which device is preferred by your organisation
Please answer the Poll and feel free to provide comments.( If I have done it correctly !!) Thanks
|
|
|
|
|
Joined: Nov 2000
Posts: 71 Likes: 16
Scholar
|
Scholar
Joined: Nov 2000
Posts: 71 Likes: 16 |
We used to use the Genius 20+ years ago and they were fine to calibrate if they were slightly out of spec. We then moved to Braun PRO4000 and in then onto PRO6000. We have the test equipment and can verify the devices and any out of spec go back to Welch Allyn under our agreement.
A couple of years back during Covid we had to start using the Genius 3. We had some test equipment from the company but even in a small room at a stable temperature we had so many calibration errors they were taking up far too much of our time on this site. The requirements for calibration were changed to something impractical, possibly quarterly, at which point they were all swapped out for PRO6000.
Solihull Hospital University Hospitals Birmingham
|
|
1 member likes this:
Geoff Hannis |
|
|
|
|
Joined: Oct 2006
Posts: 397 Likes: 11
Sage
|
Sage
Joined: Oct 2006
Posts: 397 Likes: 11 |
We have just gone over to Braun 6000 from Genius 3.
|
|
|
|
|
Joined: May 2014
Posts: 38 Likes: 9
Visionary
|
Visionary
Joined: May 2014
Posts: 38 Likes: 9 |
We have the Braun 6000.
They are used in two different ways - One is tethered to a Welch Allyn CSM and uses rechargeable batteries, the other is used as a stand alone device and uses alkaline batteries.
We get more issues with the tethers being cut and the batteries getting mixed up than the device itself!
|
|
|
|