Hi all, from the last two posts, yes the PM procedure is carried out 30-40 minutes using Datrend test kit.
With regards to the UFP (Bovie) output. On the FX8-C (Old version), and EZ-C there is a pencil output with activation for Cut and Coag, and a Bovie (UFP) output. The pencil output (the left hand tine) is the hot (active) output, but there is, in the FX8 also a UFP socket in the pencil output.
How many engineers move the footswitch to control output socket 2 no the FX8 and EZ-C and measure output on socket 2?
One of my customers had to repair one the UFP ports, but there is not an instruction to test it, it is an alternative output for the surgeon. On the EZ-C there is a second UFP only output socket, driven by a front panel footswitch, yet it does not say to test that either... Interesting. Now we have an extended PM procedure that includes these tests, available to all, free.
I note on the new FX8-FT10, there are also UFP sockets, they are even easier to test than the old FX8 and EZ-C, but note Medtronic specify a 10Mhz bandwidth for the tester (coil and RMS meter, oscilloscope, or ESU analyser). (page 6-9). The test sequences are built into the ESU, but testing takes around the same time.
More sockets on an ESU - More tests, especially cross coupling (so complicated!)
If anyone wants to discuss it further we will be the EBME show at the end of June,
Howard