Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: May 2026
Posts: 3
Janet Offline OP
Newbie
OP Offline
Newbie
Joined: May 2026
Posts: 3
Hi, I'm seeking advice on devices that test to the IEC61010 standard. Could anyone recommend any? I've been advised that all laboratory equipment must be tested to this standard. Would specialist training be required for this? Any advice greatly appreciated.

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72
I don't recall any test equipment specifically aimed at 61010. Although I suspect that even a "non-medical" PAT testing instrument may suffice.

I always used to use whatever electrical test equipment we had to hand. Any competent biomed tech should find no difficulty in carrying out EST on lab equipment. As well as inspecting the equipment with regard to any other requirements of the Standard. Just the usual common sense stuff about equipment condition and safety of use.

Hint:- pay special attention to earthing. That was always where most likely failures occurred. That and mechanical condition (there used to be a lot of tatty old items in hospital labs back in the day ... some of it even locally manufactured).

To my mind the main question is more along the lines of:- when and how often to test? Do you follow a PM regime?


If you don't inspect ... don't expect.
Joined: May 2026
Posts: 3
Janet Offline OP
Newbie
OP Offline
Newbie
Joined: May 2026
Posts: 3
Hi yes annual PM for all lab equipment however looking at the maintenance certificates they just say 'Pat test sticker ok' and make comments on a visual inspection only.
Our embe department have told us we have use the 61010 standard for all our lab equipment. I'm struggling to source a provider so we're looking into training up ourselves....

Thank you for your response smile

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72
Don't the EBME Department want to look after your equipment, then? But if you have large labs, it might be worthwhile having your own maintenance technician if you can justify such a position.

In places I have worked (mainly overseas) I (we) always included the laboratories within our remit (as well as most other technical stuff within the hospital(s)). In the larger hospitals we generally had a "lab equipment specialist" on the team.

In fact I used to enjoy working in the labs myself ... they offered a bit of respite one or two steps back from any patient environment ... as well as presenting a lot of interesting kit (not to mention clued-up people). The labs always formed the largest department by numbers of items.

What happens about the other bits of "lab equipment" that may be scattered about the hospital? Things like the Blood Gas Analyzer often hidden away in the ICU, for instance? Do the biomeds take care of stuff like that?

But annual PM ... who has been doing it? What actually gets done? What procedures are being followed?

Another thing that comes to mind is that some of the specialised stuff is (was) often covered by service contracts with outside suppliers. What do they say about 61010, I wonder?

Lastly, remember that documentation (record keeping) is one of the key factors in equipment management. And one of the most tangible benefits of following a PM schedule (but one that is often forgotten) is that it brings the technician into contact with each item of equipment on a regular basis ... a "stitch in time" and good stuff like that.


If you don't inspect ... don't expect.
Joined: Feb 2007
Posts: 302
Likes: 16
Master
Offline
Master
Joined: Feb 2007
Posts: 302
Likes: 16
The electrical safety requirements of IEC 61010 are aligned with testing standards such as EN 50699, which in turn are generally aligned with the IET Code of Practice for In-service Inspection and Testing of Electrical Equipment. So, you could use a standard electrical appliance tester or one designed for medical electrical equipment (IEC 62353).

Joined: May 2026
Posts: 3
Janet Offline OP
Newbie
OP Offline
Newbie
Joined: May 2026
Posts: 3
Thank you for your replies smile

@Geoff, I've asked a couple of service providers and they either don't do PAT testing (its an end-user responsibility) or they don't do it to IEC61010 (even the centrifuge maintenance suppliers).

I'll make a noted to ask about our Blood Gas/other POCT equipment too. Thank you!

Hmm, so I can use a medical device PAT tesing device so long as I can manually adjust or program the specific pass/fail limits dictated by the IEC 61010 standard?

Thank you for your replies, its really appreciated.

Joined: Feb 2007
Posts: 302
Likes: 16
Master
Offline
Master
Joined: Feb 2007
Posts: 302
Likes: 16
The tests of IEC 61010 are type tests, so not intended for in-service testing. You can only test using BS EN 50699 or the IET CoP and use the limits from IEC 61010, which are really the same as for other electrical appliances, such as AV or IT equipment.

If a service provider is not undertaking an electrical safety test how do they know the equipment is safe? If someone was injured (e.g., by an electric shock or burn) after their visit the HSE would have a field day with them!

Joined: May 2022
Posts: 8
Likes: 2
Newbie
Online Content
Newbie
Joined: May 2022
Posts: 8
Likes: 2
There is some really good advice in this thread, specifically what Mike mentions above is spot on and how a number of organisations approach this.

Also it might be worth reaching out to other departments/trusts to see what they do because if you are having these conversations no doubt others would have too and they maybe able to provide some insight I'm pretty sure devices like autoclaves are covered under 61010 so are there any in your trust? Maybe talk to estates or your local sterilisation hub if not.

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72
Somewhat tangential to the original questions, perhaps ... but here goes:-

Historically speaking, in British hospitals there have always been some departments that have generally "done their own thing" when it comes to technical support of equipment.

The laboratories being the example being discussed here ... but others may (will) include radiology, maybe CSSD and perhaps physiotherapy and even theatres as well (I have comes across many such examples over the years).

In my opinion (often proffered on here in the past) is that technical support of all equipment in any hospital should be the responsibility of a single office. For example:- the Hospital Engineer; with the Clinical Engineer (Biomedical Engineer or whomever) likely being the most suitable person to hold such a position.

In that way some consistency may be applied "across the board" to engineering issues, the management of service contracts, maintenance and the support of equipment generally.

PS: I notice that Janet is in the Isle of Man. Do any special rules apply there, I wonder? In other words, which organisation or authority holds sway (jurisdiction, whatever)?


If you don't inspect ... don't expect.
Joined: May 2022
Posts: 8
Likes: 2
Newbie
Online Content
Newbie
Joined: May 2022
Posts: 8
Likes: 2
Originally Posted by Geoff Hannis
Somewhat tangential to the original questions, perhaps ... but here goes:-

In my opinion (often proffered on here in the past) is that technical support of all equipment in any hospital should be the responsibility of a single office. For example:- the Hospital Engineer; with the Clinical Engineer (Biomedical Engineer or whomever) likely being the most suitable person to hold such a position.

That's the dream Geoff! smile In an ideal world, however It's one thing I find fascinating seeing how different hospitals do different things and why.

Page 1 of 2 1 2

Moderated by  Huw 

Link Copied to Clipboard
Who's Online Now
3 members (Geoff Hannis, BiomedR, Neoteny), 5,052 guests, and 22 robots.
Key: Admin, Global Mod, Mod
Newest Members
Tekie 77, Szymon1990, Janet, james_on_ebme, Biomed Egypt
10,366 Registered Users
Forum Statistics
Forums26
Topics11,256
Posts74,517
Members10,366
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5