What makes you think that the response time of the pump is the problem? Does it not just sit there, patiently waiting for signals, ready to respond immediately?
As you probably know, you can prove the pump's RS-232 interface by carrying out the "Comms test" (code 123) - and linking pins 2 and 3 (Tx and Rx),
etc.Maybe the "problem" is at the other end. The settings at the PC com port. What set-up are you using? Serial cable, PC
etc.?
What does the "Protocol" say about Baud rates (if anything)?