Hi Geoff.

Thanks for the kind welcome & quick response. I have a copy of the Alaris User manual, the Technician manual and the Asena Comms Protocol. It seems fairly straight forward really, but I just can't get it to work. I've read this thread in full, it's been really useful.

Cablewise, I have a normal USB to DB9 serial port adaptor cable, which is cabled directly to the back of the Alaris pump. I'm using the settings from the User manual:

38400 baud
1 start
1 stop
no parity
8 data bits

and I'm not sure about flow control. In the manual it says there's no Hardware flow control, so that's no to RTS/CTS I assume. I've been trying with 'none', but also with XON/XOFF.

So far I've been able to work out how to get the pump into technician mode. Then I've used code XXX so I can enable the Nurse Call and RS232 Hardware. So I assume that I've got a version of software which will allow this. So then I fire up a serial terminal program (I'm using Ubuntu Linux 13.10) like Kermit or GTKTerm. Paste the command in:

21 49 4E 46 7C 34 39 36 31 0D

But nothing happens. No response. Not even some garbled messages. Ideally I'd like to use the SYRFORCE command, which would be:

!SYRFORCE|1E6F

and in Hex:

21 53 59 52 46 4f 52 43 45 7c 31 45 36 46 0D

I'm sure once I've got the INF command working, then getting the other ones going will be fine. I don't want to do the remote config, just poll the device for some basic info.

Thanks in advance, Matt