Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#34138 02/10/08 2:36 PM
Joined: Sep 2000
Posts: 50
Scholar
OP Offline
Scholar
Joined: Sep 2000
Posts: 50
Is anyone familiar with the Viasys SiPAP (neonatal bi-level nasal CPAP)I have an error 33 which I supect might relate to Oxygen calibration required, but we have yet to obtain manuals/ training as very new. I am awaiting a response from Viasys, but any info on fault codes would be useful.
Thanks.

Last edited by John Armour; 02/10/08 2:37 PM.
Joined: Jan 2006
Posts: 57
Scholar
Offline
Scholar
Joined: Jan 2006
Posts: 57
Hi John,
We had these i the NICU/SCBU at Leigton Hospital In Crewe. I never encountered an error 33, plenty of 51,53 which relates to the O2 Cell.
If your really stuck I know the EBME guys there have a manual, I am sure if you give them a call they will check the manual for you.
If you ask for Andrew or Mark, they are very helpful guys.

Direct Dial: 01270 612234

Best Regards

Dom

Joined: Sep 2008
Posts: 2
Newbie
Offline
Newbie
Joined: Sep 2008
Posts: 2
According to the manual "E33 - Unusable - pressure sensor readings unreliable" Corrective action "Replace pneumatics PCB and recalibrate pressure and 02 sensors".

We also had problems with o2 fuel cells (error E50,E51) and changed from analytical to maxtec. Hope this helps.

Joined: Nov 2005
Posts: 160
Likes: 4
Mentor
Offline
Mentor
Joined: Nov 2005
Posts: 160
Likes: 4
Hi John

When Viasys come to look at the SiPAP they may be able to re-calibrate it. I have had this fault before and after calibrating the pressure, the fault has not surfaced again.
Saying that, if it is new and under waranty, it may be worth trying to get another board, just incase it happens again outside of warranty.

Mark


Mark Radbourne
I work for Löwenstein Medical in the UK
Joined: Sep 2000
Posts: 50
Scholar
OP Offline
Scholar
Joined: Sep 2000
Posts: 50
Thanks for you comments , everybody, sounds like I need to call them out to sort it.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 2,888 guests, and 22 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,250
Posts74,485
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5