Just wondering how many hospitals in the UK have these pumps and what issues you are having with them if any?

Joined: Jul 2000
Posts: 1,968
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,968
Likes: 32
We recently purchased the 7000 infusion pumps. Early days, but staff like them. The syringe pump has the same user interface, and we are just about to purchase over 100 (waiting final approval). I will let you know once we have had the a few months. smile


Be Proactive and reactive.
Joined: Jun 2001
Posts: 464
Sage
Offline
Sage
Joined: Jun 2001
Posts: 464
Hi John,

We only have the one 6000 syringe driver, so I can't comment on its reliability.
Arcomedical are a proactive company - any issues that you have with the 7000 series will be addressed.

Cheers
Mark

Last edited by Mark.N; 04/10/08 1:36 AM.
Joined: Jan 2006
Posts: 57
Scholar
Offline
Scholar
Joined: Jan 2006
Posts: 57
Hi Guys,
We used to have a mix of VP6000 & VP7000. If you are looking for some reliability data. Then contact Huffy on these forums. You may see some posts from him related to these devices.

Best Regards

Dom

Joined: Sep 2008
Posts: 31
Visionary
Offline
Visionary
Joined: Sep 2008
Posts: 31
Mark
What evidence are you working on to make the statement that "Arcomed are a proactive company" would a proactive company close down for 2 weeks in october with no technical or spare support?

Joined: Sep 2008
Posts: 31
Visionary
Offline
Visionary
Joined: Sep 2008
Posts: 31
Here is a test for all SP 6000 users.

Load a 5ml syringe and simulate a 2ml infusion @ 75ml/hr. Now without turning the pump off insert a 50ml syringe and attempt a 2ml @ 200ml/hr infusion.

Joined: Oct 2007
Posts: 3
Newbie
Offline
Newbie
Joined: Oct 2007
Posts: 3
Originally Posted By: Charlie A
Here is a test for all SP 6000 users.

Load a 5ml syringe and simulate a 2ml infusion @ 75ml/hr. Now without turning the pump off insert a 50ml syringe and attempt a 2ml @ 200ml/hr infusion.


The syramed pumps have multiple safety features to limit the maximum rate. During an infusion the maximum rate is limited to the lowest constraint during that infusion.

Constraints can be the syringe size but also max soft and hard limits programmed and loaded through the selected drug or limits programmed through the service mode. If you had a specific drug that limits the rate to say 50 ml/h the pump will keep that limit to 50ml/h even if you put in a larger syringe. The same obviously happens if a small syringe is loaded first and later on a larger syringe.

In the clinical practice we recommend to switch off the pump completely before starting a NEW infusion. We consider that when a syringe with a different size is loaded this will be another infusion.

Note that you don't have to switch off the pump when just loading a new syringe of the same size for the same patient with the same drug.

Joined: Oct 2007
Posts: 3
Newbie
Offline
Newbie
Joined: Oct 2007
Posts: 3
Originally Posted By: Charlie A
Mark
What evidence are you working on to make the statement that "Arcomed are a proactive company" would a proactive company close down for 2 weeks in october with no technical or spare support?


Being a proactive company we advise our customers well in advance to minimise disruption smirk.

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72
'Tis indeed a rare pleasure to see you on the forum, Cliff. smile


If you don't inspect ... don't expect.
Joined: Jul 2007
Posts: 18
SLR Offline OP
Novice
OP Offline
Novice
Joined: Jul 2007
Posts: 18
What about rogue errors? F05, F02 and battery indicator? Any issues with these?

Page 1 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,423 guests, and 291 robots.
Key: Admin, Global Mod, Mod
Newest Members
Mark B, Royal, davidarhui, Czr, Jimmy Li
10,391 Registered Users
Forum Statistics
Forums26
Topics11,279
Posts74,591
Members10,391
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5