Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#4566 24/03/06 2:13 AM
Joined: Dec 2002
Posts: 14
Novice
OP Offline
Novice
Joined: Dec 2002
Posts: 14
Have any of you come across an error code 159 or error code 161 on this Valley lab machine. It seems to put this error out the moment the settings are set beyond 10 in both cut and coag mode.The error codes are not mentioned in the service manual.159 code comes on cutting mode and 161 comes on in coag mode.

Thanks
Manoj

#4567 24/03/06 9:04 AM
Joined: Jun 2005
Posts: 24
Novice
Offline
Novice
Joined: Jun 2005
Posts: 24
Error 161 is a component failure, usually the large MOSFET in the RF output circuit (mounted to the right-hand heat sink)OR the cut/coag relay has overheated and melted. Once the error has gone check the waveform at TP19 as sometimes the caps and inductors in this circuit need replacing after this fault.

Valleylab charge aroung £130 for the MOSFET, so it may be worth looking for a local supplier.

Never seen Error 159, but it could be related to this fault.

Hope this helps!


Nil Satis Nisi Optimum
#4568 03/04/06 7:41 AM
Joined: Dec 2002
Posts: 14
Novice
OP Offline
Novice
Joined: Dec 2002
Posts: 14
Thanks StuC, I did replace the large MOSFET from another unit which was behaving erratically and the the cut/coag relay actually melted from the overheat. So had to replace both the MOSFET and the relay . The MOSFET costs A$150.00.from Tyco Australia.

Cheers
Manoj


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 2,073 guests, and 3,616 robots.
Key: Admin, Global Mod, Mod
Newest Members
Abdulaziz abed, Fernando Henz, Nick.A, Tdayve, AdrianT82
10,373 Registered Users
Forum Statistics
Forums26
Topics11,262
Posts74,527
Members10,373
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5