Hi,

I have to repair a defibrillator cardio Aid MC + (yellow) which appears in the dispaly the message "DEFI out of order - do not use (2)". Can anyone help me?

Best regards

Mofe #56756 08/06/11 1:49 PM
Joined: Nov 2010
Posts: 60
Scholar
Offline
Scholar
Joined: Nov 2010
Posts: 60
Hi
If the error message "DEFI FAULT -DO NOT USE [2], it means that charge timed out, faults detected in the HV circuit during charge or ready phase
SJ with best regard

Mofe #57092 04/07/11 4:51 PM
Joined: Jul 2011
Posts: 71
Likes: 11
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 71
Likes: 11
I have service manual in paper. I see for diagnostics.

Joined: Nov 2011
Posts: 2
Newbie
Offline
Newbie
Joined: Nov 2011
Posts: 2
Hi!

I have a problem... I bought a CardioAid Mc+ for my ambulance. It works quite good, but I'm not able to set the date. Now the date is "04/06/2046" and the time is "30:33"... Is it possible?? I have no service manual and I don't know what to do. No one in my city repairs this model of defib. Could you help me? Thank you very much!!


Last edited by Croce Blu Italia; 15/11/11 12:50 AM.

Tiziana
Croce Blu Italia
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Welcome to the forum, Tiziana. smile

For some reason I have generally found information available on the 'net for *S&W equipment to be in short supply.

Also, I'm not so sure that the Cardio-Aid MC+ sold in great numbers here in the UK.

But hopefully, someone on here will come up with the information you need.

In passing, I would comment that it sounds to me like a battery is on the way out (I'm guessing at a button cell on a circuit board somewhere).

Meanwhile, I'm interested the Italian Blue Cross insignia. Is that EU iconography I see? think

* Other labelling may have been used:- Artema for example.


If you don't inspect ... don't expect.
Mofe #59104 15/11/11 12:56 PM
Joined: Oct 2003
Posts: 51
Scholar
Offline
Scholar
Joined: Oct 2003
Posts: 51
From the Service manual

28.2 Real-time clock
a) Switch on the defibrillator.
b) Check the correctness of the day, month, year, hours,
minutes and seconds shown.
c) If adjustments are needed, enter the configuration
menu by pressing Freeze during the self test period
at power on.

29.2 Real-time clock
a) Switch on the defibrillator.
b) Check the correctness of the day, month, year, hours,
minutes and seconds shown.
c) If adjustments are needed, enter the configuration menu
by pressing the second softkey from the left during the
self test period at power on.

Joined: Nov 2011
Posts: 2
Newbie
Offline
Newbie
Joined: Nov 2011
Posts: 2
@Geoff Hannis: Thank you for the information about the CardioAid MC+.

Originally Posted By: Geoff Hannis
Meanwhile, I'm interested the Italian Blue Cross insignia. Is that EU iconography I see? think

Usually in Italy ambulance services are called "cross" + a color (not only Red Cross, but also Blue Cross, White Cross, Green Cross, etc). Everyone can create his own logo, if not already registered by others. We have choosen a blue cross, with the colors of the italian flag all around (red, white, green). In the middle of the cross there are the twelve yellow stars of EU.


@Paul Robbins: Thanks to have posted the extract from the service manual! Friday I will be in Biella again and I'll try to do as you suggested. smile


Tiziana
Croce Blu Italia
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

I hope you get it sorted out OK (it should not be a problem).

Meanwhile (as a "numbers man") I find it noteworthy that the EU finally settled on twelve stars.

That is, a good old British dozen ... rather a Continental ten (or any other number). think

By the way, in the UK the Blue Cross is a long-established Charity looking after unwanted (and abused etc.) animals.


If you don't inspect ... don't expect.
Mofe #63770 12/02/13 8:35 AM
Joined: Mar 2011
Posts: 10
Novice
Offline
Novice
Joined: Mar 2011
Posts: 10
S&W Cardio Aid MC+ "Calibration Needed" How do I solve this problem?
Thanks..

Mofe #63774 12/02/13 1:19 PM
Joined: Mar 2011
Posts: 10
Novice
Offline
Novice
Joined: Mar 2011
Posts: 10
and what does it mean , "TEST IS ACTIVE AT S18"

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 158 guests, and 3 robots.
Key: Admin, Global Mod, Mod
Newest Members
Royal, davidarhui, Czr, Jimmy Li, AndrewS1
10,390 Registered Users
Forum Statistics
Forums26
Topics11,276
Posts74,584
Members10,390
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5