Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#61312 12/06/12 4:50 PM
Joined: May 2012
Posts: 32
Visionary
OP Offline
Visionary
Joined: May 2012
Posts: 32
Guys need a bit of a hand with this one, this concerns the gold series and predecessors. When calibrating with the precision syringe, calibration is successful on the display but on the bellows driven chart I get an inconsistent increase, i.e. 1litre fine, then 2.1 increasing up to 7.5 for 7litres input. I've moved the base plate as advised in the tech manual, and ones fitted with a calibration scale of adjustable screws I've adjusted to no avail. I've even changed the bellows to no effect, this is common to all vitalographs I've inherited across my region, and having had no training on these, well any advice?

Joined: May 2007
Posts: 8
Newbie
Offline
Newbie
Joined: May 2007
Posts: 8
Have you spoke to vitalgraphs service department as they will be able to give you some useful information on possible solutions. They are very helpful as it's been a long time since I've worked on a gold standard machine.

Joined: May 2012
Posts: 32
Visionary
OP Offline
Visionary
Joined: May 2012
Posts: 32
Cheers mate,yes I did, they said the cams will have to be adjusted, they did not elaborate on this and then quoted 300 quid per machine to do this, I am now in the process of having a chat with my bosses about just acquiring a desk jet printer and associated cabling for about 100 quid and ignoring the bellows, printing all results and calibration reports to a desk jet.

Joined: Dec 2003
Posts: 282
Master
Offline
Master
Joined: Dec 2003
Posts: 282
You will need to remove the back then you will see about 8 hex screws or small bolts on the cam. You will need to adjust them so that you get the correct reading on the chart to correspond with each inflation with the cal syringe. The cam should end up with a gentle curve i.e. not big steps. This info is in the user manual, if you haven't got one, search Vitalographs web or phone them and see if they have a spare copy


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Czr), 5,149 guests, and 50 robots.
Key: Admin, Global Mod, Mod
Newest Members
Royal, davidarhui, Czr, Jimmy Li, AndrewS1
10,390 Registered Users
Forum Statistics
Forums26
Topics11,278
Posts74,588
Members10,390
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5