Joined: Oct 2002
Posts: 130
EH Offline OP
Expert
OP Offline
Expert
Joined: Oct 2002
Posts: 130
Hi, Does anyone know of any "Shock" sensor, or otherwise, fitted to these pumps as we have had a number been brought to the workshop constantly alarming with a "Failure" message. As usual "It just happened on its own accord" rolleyes . We have a suspicion that they are being dropped. Temporary disconnection of the battery seems to reset the pump. A thorough bench test does not show any unusual occurrences and the device passes all manufacturer tests. Any help will be most grateful, thank you. confused


A problem is a gift to be unwrapped :rolleyes:
Joined: May 2003
Posts: 15
Novice
Offline
Novice
Joined: May 2003
Posts: 15
We have approx 120 colleague pumps. This constant alarm is a constant problem. Baxter blame it on flat batteries but we have had this occur with charged batteries.
Unfortunately by disconnecting the batteries you lose the event history, so you get go and look to see what caused the alarm in the first place.
Our pumps are leased at the moment so we don't have to pay for repairs.
Apparently the Version 4 pumps are a lot better.

Joined: Oct 2003
Posts: 38
Visionary
Offline
Visionary
Joined: Oct 2003
Posts: 38
I have seen this problem with the Baxter Colleague Pump. The pump alarms and the processor seems to 'hang'. The only fault that I found was fluid ingress on the display PCB in the pump module. There is a rubber gasket that should prevent fluid ingress onto the board but if this is damaged in any way it leads to constant alarm problems.

Joined: Jul 2000
Posts: 1,965
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,965
Likes: 32
I haven't seen this problem on the colleagues we use in our oncology suite, but they are newer the version, and the oncology nurses really like them. I was speaking to the equipment library manager from Hillingdon last week and asked for his opinion, he said they were very good, reliable pumps, so looks like Baxter have got their act together with the new version pumps.
smile


Be Proactive and reactive.
Joined: Oct 2002
Posts: 130
EH Offline OP
Expert
OP Offline
Expert
Joined: Oct 2002
Posts: 130
Thanks Guys, I can now confirm that at least two of these pumps "hit the deck" so to speak. frown


A problem is a gift to be unwrapped :rolleyes:
Joined: May 2001
Posts: 193
Mentor
Offline
Mentor
Joined: May 2001
Posts: 193
John,

Do you have a contact name or e-mail for that Library manager at Hillingdon?

I have been trying to make contact fro a while now with little success. I did send an e-mail I got from a registered user on this site but no success.

Chris-H


Ours is not to reason why?,
Simply obey & then comply !
Joined: Mar 2003
Posts: 23
Dreamer
Offline
Dreamer
Joined: Mar 2003
Posts: 23
Hello Chris

We have the Colleague and so far so good, no major faults to report. Although the manual release tab has caused a 'few' user errors.

Joined: Sep 2002
Posts: 13
JET Offline
Novice
Offline
Novice
Joined: Sep 2002
Posts: 13
Hi

So were not the only ones to suffer with this problem, with the manual tube release and spurious alarms. I have found that this seems to be linked to the battery status and depends on the number of discharges below threshold and discharge cycles that have occured, the higher these numbers are the worst the problem seem to be. Are the batteries that they use good enough for the pump, we seem to suffer more with our tripples than with our singles.

John laugh

Joined: Jul 2000
Posts: 1,965
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,965
Likes: 32
Chris,
David Marshall is the contact - but you probably already spotted that. laugh


Be Proactive and reactive.
Joined: May 2001
Posts: 193
Mentor
Offline
Mentor
Joined: May 2001
Posts: 193
David
Not sure if you got my message a while back. Can you drop me an email sometime as I think the link was a null one !

Cheers
smile
Chris-H


Ours is not to reason why?,
Simply obey & then comply !

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 3,466 guests, and 26 robots.
Key: Admin, Global Mod, Mod
Newest Members
j9_PLC, nece, Vitya, Shenzhen007, Eng. Craig
10,357 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,357
Most Online37,242
Apr 12th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5