I have an Evita XL ventilator that powers up, curiously it doesn't shoe the bar loading boot image, rather it shows a Samsung logo. Then it goes straight into ventilation GUI, it bypasses the part where you chose what patient you want to ventilate. At the top it says "Loss of Data" then after about a min or two it reboots itself and repeats the cycle over and over again. It gives no error code.

Anybody know what could cause this?

Thanks

Last edited by ebmeNeophite; 18/03/15 8:59 PM.
Joined: Oct 2013
Posts: 43
Technologist
OP Offline
Technologist
Joined: Oct 2013
Posts: 43
So I found the solution, just in case someone else needs it, it's the battery on the 2nd card in the back, counting from bottom to top. It's the CPU PCB card, it has a battery and it was dead.

I just replaced it and everything is now fine and dandy.

Last edited by ebmeNeophite; 18/03/15 11:05 PM.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Well done. And thanks for letting us know. smile


If you don't inspect ... don't expect.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 11,743 guests, and 271 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,282
Posts74,585
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5