Hello EBME people.
I am in dire need of a venturi test tool, which would assist me to create a venturi effect across a patient inspiratory valve. So I can measure the lift off pressure? I have tried a few home made creations, but they are not consistent, so over to the Guru's who inhabit these forums. I have asked the miriad of companies who normally plague our workshop, but only one positive response so far. I would expect this to be a common problem with those who service ventillators etc?
Ideally able to create a negative pressure of no more than 10 mBar; as the lift off is usually less than 3 Mbar
any idea's ?
best regards
Mark Edmondson Acute ICU
|
|
|
|
Joined: Jul 2011
Posts: 65 Likes: 10
Scholar
|
Scholar
Joined: Jul 2011
Posts: 65 Likes: 10 |
If I remember correctly, the old Sechrist IV respirator had a jet with a venturi tube in the exhalation block, used to compensate the tubing flow resistance; with standard connection at the inlet and a performance of a few millibar, so by adjusting the jet flow you can have a stable and adjustable vacuum flow as needed.
|
|
|
|
Joined: Feb 2004
Posts: 14,767 Likes: 70
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,767 Likes: 70 |
Have you got something able to read such a low pressure with precision?
Have your home grown efforts used a syringe?
I'm imagining a (largish) syringe teed off to a gauge (or perhaps an electronic pressure meter) from suitable adaptor(s) to suit the patient valve. You should be able to simulate patient inhalation in such a fashion.
I seem to recall using a bellows to simulate a patient "pulling" (taking a breath) somewhere in the dim and distant past.
|
|
|
|
Joined: Jul 2011
Posts: 65 Likes: 10
Scholar
|
Scholar
Joined: Jul 2011
Posts: 65 Likes: 10 |
A differential manometer, made from a U-shaped tube or hose filled with water, is a primitive but faithful instrument, as much as a millibar is one centimeter of water. I understand the need to use a venturi to maintain a continuous and stable flow to detect the threshold of change of state of the valve.
|
|
|
|
Joined: Sep 2019
Posts: 4 Likes: 1
Newbie
|
Newbie
Joined: Sep 2019
Posts: 4 Likes: 1 |
Hi Mark, I recommend you consider the Injector supplied by Draeger. Part No 7900930. We use this device to achieve a -4 mBar pressure for testing the Emergency Respiratory Valve. In combination with a Pipeline Reducing Regulator (speak to Oxylitre for this), you are able to adjust to demand to suit your needs. Danny
|
|
|
|
Joined: Mar 2001
Posts: 273
Master
|
OP
Master
Joined: Mar 2001
Posts: 273 |
Yes Geoff I have one of these old Ohemda test lungs, I use to use it testing the Anaesthetic machines, but I cant hold the negative draw long enough to get a viable reading. on the inspiratory valve.
Daniel, because like everyone else in the NHS funds are not available I was looking for a cheaper option than using the £800 drager service tool.
VitaPablo you seemed to have grasped the problem exactly how I see it ..
The solution we found by using a wall mounted vacuum low suction set at a negative draw of 12mmHg. this is the same figure as generated off a dedicated borrowed venturri. ( when a flow between 60-65 litres used.)Solved by my trusty apprentice!
The trainer did state at the time I was trained, that a vacuum controller, could be used but they do not have that facility themselves; and one forgets.
thanks to all for your inputs
|
|
|
|
Joined: Feb 2004
Posts: 14,767 Likes: 70
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,767 Likes: 70 |
|
|
|
|
Joined: Jun 2024
Posts: 1
Newbie
|
Newbie
Joined: Jun 2024
Posts: 1 |
Please i am in Manchester and in dire need of a place I can volunteer to learn on the job, I am a biomedical Engineer with a master degree in Biomedical Engineering. Please help
Kingsdave
|
|
|
0 members (),
2,016
guests, and
10
robots. |
Key:
Admin,
Global Mod,
Mod
|
|
Forums25
Topics11,164
Posts74,193
Members10,259
|
Most Online5,980 Jan 29th, 2020
|
|
|
|
|