Hello EBME people.

I am in dire need of a venturi test tool, which would assist me to create a venturi effect across a patient inspiratory valve. So I can measure the lift off pressure?
I have tried a few home made creations, but they are not consistent, so over to the Guru's who inhabit these forums.
I have asked the miriad of companies who normally plague our workshop, but only one positive response so far. I would expect this to be a common problem with those who service ventillators etc?

Ideally able to create a negative pressure of no more than 10 mBar; as the lift off is usually less than 3 Mbar

any idea's ?

best regards

Mark Edmondson
Acute ICU

Mark.E #77921 18/07/24 12:43 PM
Joined: Jul 2011
Posts: 71
Likes: 11
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 71
Likes: 11
If I remember correctly, the old Sechrist IV respirator had a jet with a venturi tube in the exhalation block, used to compensate the tubing flow resistance; with standard connection at the inlet and a performance of a few millibar, so by adjusting the jet flow you can have a stable and adjustable vacuum flow as needed.

Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72
Have you got something able to read such a low pressure with precision?

Have your home grown efforts used a syringe?

I'm imagining a (largish) syringe teed off to a gauge (or perhaps an electronic pressure meter) from suitable adaptor(s) to suit the patient valve. You should be able to simulate patient inhalation in such a fashion.

I seem to recall using a bellows to simulate a patient "pulling" (taking a breath) somewhere in the dim and distant past.


If you don't inspect ... don't expect.
Mark.E #77929 22/07/24 10:53 AM
Joined: Jul 2011
Posts: 71
Likes: 11
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 71
Likes: 11
A differential manometer, made from a U-shaped tube or hose filled with water, is a primitive but faithful instrument, as much as a millibar is one centimeter of water.
I understand the need to use a venturi to maintain a continuous and stable flow to detect the threshold of change of state of the valve.

Mark.E #77932 24/07/24 12:44 PM
Joined: Sep 2019
Posts: 4
Likes: 1
Newbie
Offline
Newbie
Joined: Sep 2019
Posts: 4
Likes: 1
Hi Mark,
I recommend you consider the Injector supplied by Draeger. Part No 7900930.
We use this device to achieve a -4 mBar pressure for testing the Emergency Respiratory Valve.
In combination with a Pipeline Reducing Regulator (speak to Oxylitre for this), you are able to adjust to demand to suit your needs.
Danny

Joined: Mar 2001
Posts: 273
Mark.E Offline OP
Master
OP Offline
Master
Joined: Mar 2001
Posts: 273
Yes Geoff I have one of these old Ohemda test lungs, I use to use it testing the Anaesthetic machines, but I cant hold the negative draw long enough to get a viable reading. on the inspiratory valve.

Daniel, because like everyone else in the NHS funds are not available I was looking for a cheaper option than using the £800 drager service tool.

VitaPablo you seemed to have grasped the problem exactly how I see it ..

The solution we found by using a wall mounted vacuum low suction set at a negative draw of 12mmHg. this is the same figure as generated off a dedicated borrowed venturri. ( when a flow between 60-65 litres used.)Solved by my trusty apprentice!

The trainer did state at the time I was trained, that a vacuum controller, could be used but they do not have that facility themselves; and one forgets.

thanks to all for your inputs

Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72
Promote that apprentice!


If you don't inspect ... don't expect.
Joined: Jun 2024
Posts: 1
Newbie
Offline
Newbie
Joined: Jun 2024
Posts: 1
Please i am in Manchester and in dire need of a place I can volunteer to learn on the job, I am a biomedical Engineer with a master degree in Biomedical Engineering. Please help


Kingsdave

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Huw), 1,799 guests, and 58 robots.
Key: Admin, Global Mod, Mod
Newest Members
Nick.A, Tdayve, AdrianT82, Iagonuns, Dctm
10,371 Registered Users
Forum Statistics
Forums26
Topics11,259
Posts74,523
Members10,371
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5