Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
#77716 26/02/24 1:05 PM
Joined: Nov 2000
Posts: 72
Likes: 16
Scholar
OP Offline
Scholar
Joined: Nov 2000
Posts: 72
Likes: 16
We are currently using Draeger JM-105 Jaundice meters for our community midwives. These units are not cheap, the last ones purchased were just under £4k each. I'm searching for alternatives to recommend to the staff, can anyone recommend an alternative brand for us to look at? Also, any info on on-going maintenance costs too if available.


Solihull Hospital
University Hospitals Birmingham
Joined: May 2014
Posts: 38
Likes: 9
Visionary
Offline
Visionary
Joined: May 2014
Posts: 38
Likes: 9
We have purchased the below:

https://www.mennenmedical.com/bilicare/

Been in nearly a year and have had no issues, unsure of on going costs.

The unit cost £2400 in vat

1 member likes this: Adrian Jones
GlynEd #78355 28/05/25 8:26 AM
Joined: Nov 2000
Posts: 72
Likes: 16
Scholar
OP Offline
Scholar
Joined: Nov 2000
Posts: 72
Likes: 16
Hi Glyn,

How are you getting on with these devices? Have there been any calibration or mechanical issues?


Solihull Hospital
University Hospitals Birmingham
Joined: Dec 2015
Posts: 163
Likes: 30
Mentor
Offline
Mentor
Joined: Dec 2015
Posts: 163
Likes: 30


30 years since the Chernobyl disaster and yet we still have no super heroes or zombies.
1 member likes this: carl ray
Dustcap #78357 03/06/25 10:01 AM
Joined: Nov 2000
Posts: 72
Likes: 16
Scholar
OP Offline
Scholar
Joined: Nov 2000
Posts: 72
Likes: 16
Hi Dustcap,

Thank for the heads-up with Walters. I found this unit with QED Scientific some time back.

I did some research on these a year or so ago and contacted importers / manufacturers.

We have a lot of Draeger JM-105 units that get sent off for calibration when needed and so looking for a device that can be calibrated in-house as and when we need to. The Bilicare unit does not require service or calibration so looks favourable. The JM-105 and BM-100A both have built in self-checkers. For the BM-100A should the values differ significantly from lab values it requires 15 samples, high and low values, that need to go through a lab. A compensation value can then be calculated and entered in to the device. For this reason we'd probably go with the Bilicare but would like user / engineer feedback.

I'll be at the Expo and will no doubt see Richard from Walters Medical so will have a chat with him as well.


Solihull Hospital
University Hospitals Birmingham
Joined: May 2014
Posts: 38
Likes: 9
Visionary
Offline
Visionary
Joined: May 2014
Posts: 38
Likes: 9
Hi Adrian,

We have 5 of the Bilicares in clinical use.

They are just over 2 years old now and we have only had one major fault, the device wouldn't switch on - no problems with the battery or charger so it got sent back to Charter Kontron for repair - cost approx £350 to rectify.

Apart from that they have been reliable and caused us no issues (apart from finding them!)

1 member likes this: Adrian Jones

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 10,783 guests, and 151 robots.
Key: Admin, Global Mod, Mod
Newest Members
Edouardohaag, Mark B, Royal, davidarhui, Czr
10,392 Registered Users
Forum Statistics
Forums26
Topics11,279
Posts74,581
Members10,391
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5