- Using an old computer or laptop with a traditional RS232 port (in stead of using USB) could make all the difference.

- Be sure to enable Remote Control and Nurse Call.

- I use a Baud Rate of 38400. I don't know if this makes a difference for you, but it works for me on this setting.

- Also, check the software version on the pump (I think you can do this via menu 123 or 251, but do check the manual to be sure). Up to version 2.3.6, remote control can be enabled. In newer versions, this option is disabled.