|
|
|
|
Joined: Jun 2004
Posts: 4
Newbie
|
Newbie
Joined: Jun 2004
Posts: 4 |
It will be fairly apparent if the plunger arm has been forced. The syringe force will be way out.
PL3 is usually down to syringe changes. When the user get a NEOI alarm, they dont stop the pump using the hold button. They just move the plunger back while the motor is still driving froward. The watchdog get confused as the plunger position reading is wildly out from the actual. So it flags up a "PL3" error.
The only thing that you can check is running the linearity check. And if thats out (>0.7mm we recal), you would calibrate the plunger position.
We have had (into the 100's) of PL3 errors. Its a training issue as they do need to stop the infusion before changing syringes.
Thanks
Robbie
|
|
|
|
|
Joined: Oct 2009
Posts: 1
Newbie
|
Newbie
Joined: Oct 2009
Posts: 1 |
George,
Could you send me the pdf of the service manual aswell please.
It would be really helpful for me, and appreciated.
Thanks, FD
|
|
|
|
|
Joined: Aug 2009
Posts: 41
Technologist
|
Technologist
Joined: Aug 2009
Posts: 41 |
PL3 failure for alaris is initiated when its main processor s/w has detected an excessive deviation of greater than 2.4mm for longer than 0.8 seconds between the expected syringe plunger position with respect to the actual syringe position.
I have had this fault occur before on one specific pump, so i sent it back to cardinal for a full inspection. If need be, you can email me and i can send you a copy of the service repot which i recieved if it will help.
Dee
|
|
|
|
|
Joined: Feb 2004
Posts: 14,804 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,804 Likes: 72 |
Did they find anything wrong? Or was it the usual problem as described above by Robbie? 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Aug 2009
Posts: 41
Technologist
|
Technologist
Joined: Aug 2009
Posts: 41 |
Contact contamination pl5 ( conector between the linear soft potentiometer and the chassis pc board )
|
|
|
|
|
Joined: Feb 2004
Posts: 14,804 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,804 Likes: 72 |
Due to? Fluid ingress? 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Jun 2010
Posts: 6
Newbie
|
Newbie
Joined: Jun 2010
Posts: 6 |
i need ur help my alaris asena GS error SP2.I refer technical manual: it's because safety processor ID version failure.this may occur if comms to the safety processor fails. what should i do??
|
|
|
|
|
Joined: May 2008
Posts: 50
Scholar
|
Scholar
Joined: May 2008
Posts: 50 |
I have seen components become detached from Asena PCBs before, so it might be worthwhile checking all the solder legs are still attached to the board. Unfortunately, I can't remember where on the main PCB the safety processor lives (not had one open for about 3 years).
Aside from that, us profligate NHS folks would tend to just replace the main PCB (or control PCB as the manufacturer calls it).
|
|
|
|
|
Joined: Sep 2010
Posts: 20
Dreamer
|
Dreamer
Joined: Sep 2010
Posts: 20 |
Any safety processor errors you should replace the main PCB as Gordovan says....
|
|
|
|
|
Joined: Apr 2011
Posts: 6
Newbie
|
Newbie
Joined: Apr 2011
Posts: 6 |
i need ur help my alaris asena GS error SP2.I refer technical manual: it's because safety processor ID version failure.this may occur if comms to the safety processor fails. what should i do?? Hy! I'm from Romania and I have the same problem to an Alaris Asena GS and I don/t know what to do.I need the manual service for future. Somebody can help me? Thanks!
|
|
|
|
1 members (MikeX),
3,160
guests, and
8
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,256
Posts74,514
Members10,366
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|
|
|