Originally Posted By: MikeX
Many commonly used test instruments on the market, even new units, that use an AC source for the test current can give incorrect readings when an inductor is present in the earth connection (as used for EMC). This is a known issue with many Fluke Biomed meters as well as the Rigel 266.


News to me!

Earth bonding , this is a low impedance path to earth to prevent hazardous voltages appearing on the equipment.
Protective Earth Conductor , conductor to be connected between the protective earth terminal and an external protective earthing system.
Protective Earth Terminal , connected to conductive parts of Class I equipment for safety purposes. This terminal is intended to be connected to an external earthing system by a protective earth conductor.
Potential Equalization Conductor , conductor providing a connection between equipment and the potential equalization busbar of the electrical installation (involves joining together metalwork that is or may be earthed so that it is at the same voltage).

In IEC60601, the test is conducted using a 50Hz current between 10A and 25A for a period of at least 5 seconds with a no load voltage not exceeding 6V, when testing the impedance of a protective earth.

Originally Posted By: MikeX
Those meters that use a DC test current, such as the Rigel 288 and GMC Secutest SIII, do not exhibit this problem.

Can you provide some evidence that these tester's use a DC Test current?

An inductor will have a negligible effect on PE readings at 50 Hz, what you are observing is related to the higher current levels overcoming contact oxidisation and providing a lower impedance measurement.