Have you tried this yourself? I bet a low current test will show up a high resistance value for a single strand as
A) the mechanical contact will have been significantly compromised if this was the case (ie providing constriction resistance) and
B) the cross section of the earth wire will have been significantly reduced thus increasing the resistance for electrical current.

I have done tests whereby the earth screw was loosened (could happen due to vibration and more likely than the single strand scenario), compromising both the mechanical and electical security of the protective earth (ie high constriction resistance for electrical current). A 25A tests gave a good pass, the low current saw a high resistance and failed. Just another scenario to consider.............