I will ask for some feedback from Ultramedic on this one regarding Fluke Biomedical USA assertion of meter circuit errors, which the likelihood of occurring (by your confirmation) is minimal.

It does not really show us anything, (as no one seriously believes that test equipment is 100% accurate or reliable anyway), most users routinely use proving units to confirm that their test equipment is performing within manufacturers specifications in between calibrations, and are aware of errors within published tolerance values. If you’re not, you shouldn’t be using test equipment.

Most competent users of safety testers checking to IEC60101 will find that the available 1A, 10 & 25A Earth bonding type tests on existing testers are quite adequate and have never caused damage.

I personally disagree; complying with best practice does not make IEC 62353 the only choice at all. DB 2006(05) is guidance only. Any conscientious Biomed when faced with two standards for Electrical Safety of Medical Equipment would utilise the higher standard for tests, as opposed to simplified tests.

In the current economic climate, I would select the lower current selection options available for PE measurement in an IEC60101 tester, and thus save the Trust & the Taxpayer the cost of buying a new safety tester that has the IEC 62353 option installed. We are all accountable for public monies.

The argument that simplified tests ensure minimum costs and maximum safety sounds like a sales pitch, but I could not justify the expense, when perfectly good existing electrical safety testing equipment is available , and can be utilised.
I can be confident that I am complying with my Trusts Standing Financial Instructions, (not to waste money).