Originally Posted By: Sean Fearon

Whilst we are considering the differences between the two standards, the IEC 60101 checks earth leakage in Single and Double Fault condition, I am unable to determine if the IEC 62353 performs the earth leakage test in Double Fault condition?


The IEC 62353 direct equipment leakage check is only tested in a single fault condition but with mains polarity in the normal and reversed conditions.

If you want to perform the double fault condition test with an open earth and open neutral you can use the IEC 62353 alternative method equipment leakage test. This will provide a result exactly the same as the open earth/open neutral IEC 60601-1 test, assuming the mains supply is 230V + 10% (as the tester uses a 253V AC source).

It is really worth while getting to know IEC 62353 as it is not as bad as some would have you believe!