EARTHBONDING TESTS (Rigel Medical Practical Guide to IEC 62353)

High Test Currents (10A or more) might potentially (my itallics) be destructive to parts of the unit under test which are connected to the protective Earth but have a functional purpose (e.g. screening). As such, consideration should be given to the test current.

Low Test Currents (<8A) could potentially influence the reading as contact resistance is influenced by a number of factors (Constriction, Pressure, Film resistance).
Higher Test Currents overcome the contact resistance where lower currents show a relatively higher reading, thus potentially causing unnecessary failures.

We still have a lot of kit in Theatres, such as stack systems with earthed chassis isolating transformers on long mains leads with multiple bonding connections which require the use of 10A and higher earth bonding tests. It would appear that deciding to go down the line of exclusive use of IEC 62353 would give rise to an increased use of Medical system PE failures (after the checking of individual devices).

Whilst we are considering the differences between the two standards, the IEC 60101 checks earth leakage in Single and Double Fault condition, I am unable to determine if the IEC 62353 performs the earth leakage test in Double Fault condition?